From owner-structural-nmr@net.bio.net Wed Jan 01 22:00:00 1997
Path: biosci!ihnp4.ucsd.edu!munnari.OZ.AU!news.ecn.uoknor.edu!feed1.news.erols.com!news-xfer.netaxs.com!netnews.upenn.edu!dellwo
From: dellwo@spruce.chem.upenn.edu (Martin J. Dellwo)
Newsgroups: bionet.structural-nmr
Subject: Re: 3D NMR processing on Felix950
Date: 2 Jan 1997 18:19:56 GMT
Organization: Department of Chemistry, University of Pennsylvania
Lines: 32
Message-ID: <5agu8c$3fb@netnews.upenn.edu>
References: <32CAA3AA.41C6@tracer.chem.uh.edu>
Reply-To: dellwo@spruce.chem.upenn.edu
NNTP-Posting-Host: spruce.chem.upenn.edu

In article <32CAA3AA.41C6@tracer.chem.uh.edu>,
Hong Yu Liu  <hong@tracer.chem.uh.edu> wrote:
>	I acquired the NOESY-HMQC data on a bruker600 AMX instrument with the
>following TD: t3 X t2 X t1: 1024 X 32 X 256. 
>	I used the conversion program supplied by felix X32 New to convert the
>bruker format to felix format. It seems to went through fine ( at least
>no complaining messages). 

>	Matrix dimension: D1 X D2 X D3: 512 X 32 X 256
>
>	Acquisition Data
>	acquisition type: Quartets
>	1st incremented: t2
>	# of t2 exps: 32
>	# of t1 exps: 256

It is highly unlikely that '1st incremented' was t2 and not t1.  You
probably should run it with t1 as the '1st incremented', and matrix
dimensions d1xd2xd2:512x256x32

Also, since you used the x32_new to convert the data, you should
verify that the transform used is the 'dft' command, with the correct
decimation factor.  If your traces looked ok after processing D1, than
this was probably the case.

I do not use the EZ transforms, so I can't tell you if the remaining
parameters were correct (specifically, I can't tell you if
'acquisition type: Quartets' is correct).
-- 
Martin J. Dellwo   (215) 898-4886      dellwo@spruce.chem.upenn.edu
Department of Chemistry, University of Pennsylvania
http://cherry.chem.upenn.edu/~dellwo/

From owner-structural-nmr@net.bio.net Thu Jan 02 22:00:00 1997
Path: biosci!SPRUCE.CHEM.UPENN.EDU!tan
From: tan@SPRUCE.CHEM.UPENN.EDU (Wee "Derrick" Tan)
Newsgroups: bionet.structural-nmr
Subject: ROESY-HSQC
Date: 3 Jan 1997 13:52:34 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 15
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701032104.QAA14159@spruce.chem.upenn.edu>
NNTP-Posting-Host: net.bio.net

Dear NMR specialists,

	Does any one of you have the pulse sequence for a 3D ROESY-HSQC or HMQC
encoded for the Bruker system?

	I will appreciated it very much if anyone of you can provide me with
one. Thanks in advance.

-- 
Derrick Tan
University of Pennsylvania	Phone    : (215) 898-4886
Department of Chemistry		Fax      : (215) 573-2123
231 S. 34th St.			Email    : tan@osage.chem.upenn.edu
Philadelphia, PA 19104		Homepage : http://cherry.chem.upenn.edu/~tan/


From owner-structural-nmr@net.bio.net Thu Jan 02 22:00:00 1997
Path: biosci!bcm.tmc.edu!cs.utexas.edu!howland.erols.net!newsxfer3.itd.umich.edu!portc01.blue.aol.com!newsstand.cit.cornell.edu!news.acsu.buffalo.edu!acsu.buffalo.edu!marcus
From: marcus@acsu.buffalo.edu (Emil Marcus)
Newsgroups: bionet.structural-nmr
Subject: BRUKER to VARIAN conversion
Date: 3 Jan 1997 18:53:22 GMT
Organization: State University of New York at Buffalo
Lines: 12
Distribution: world
Message-ID: <5ajkj2$7d4@prometheus.acsu.buffalo.edu>
NNTP-Posting-Host: autarch.acsu.buffalo.edu
NNTP-Posting-User: marcus


Dear NMR specialists:

I would be grateful if you could help me find computer programs that 
are capable to convert BRUKER (raw data) files into VARIAN files,
to be used with VNMR, which is VARIAN's own spectra processing program.

Emil

Emil Marcus
University at Buffalo
marcus@acsu.buffalo.edu

From owner-structural-nmr@net.bio.net Sat Jan 04 22:00:00 1997
Path: biosci!PHOENIX.PRINCETON.EDU!ipelczer
From: ipelczer@PHOENIX.PRINCETON.EDU (Istvan Pelczer)
Newsgroups: bionet.structural-nmr
Subject: Re: Linear Prediction vs. First Point Scaling
Date: 5 Jan 1997 09:30:58 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 52
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SUN.3.91.970105121054.8330A-100000@phoenix.princeton.edu>
References: <32CFB108.14A3@plaza.snu.ac.kr>
NNTP-Posting-Host: net.bio.net


Dear Ji Hyun,
It is the position of the first point in time which decides (at the first 
place) if there is need for correction.  Linear prediction in usuual 
circumstances does not change digitization, therefore if your experiment 
was set up with first point at half dwell time then this will be retained 
after LP replacement of some of the first points with no correction 
required.  An alternative is (giving up advantageous aliasing properties 
in the frequency domain) to detect "first point" at a full dwell time, 
then back calculate the true first point (t1=0) using LP.  In this case 
0.5 correction is needed in an ideal case.
Both alternatives were presented in the literature from Ad Bax's lab, and 
you can find some description and references in our reviews on data 
processing in multidimensional NMR (Pelczer & Szalma, Chem. Rev., 
91(1991)1507, or in an upcoming book from Humana Press, edited by D. G. 
Reid:  Pelczer & Carter, Data Processing in Multidimensional NMR, 
Chapter for the book: Protein NMR Protocols, in the series: Methods in 
Molecular Biology).
It is another issue if the first point(s) are distorted by either 
hardware or errors by LP.  Then correction by other than the theoretical 
value may be required.
All the best, and good luck with processing,

Istvan

wwww,wwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwww
Istvan Pelczer, Ph.D.		       		Email: ipelczer@princeton.edu
Senior NMR Spectroscopist
Princeton University
Department of Chemistry, Frick Lab.,	 		 ph#  (609) 258 2342
Washington Road						fax#  (609) 258 6746
Princeton,  NJ 08544, USA




On Sun, 5 Jan 1997, Lee, Ji Hyun wrote:

> To my knowledge, if we don't half dwell time delay during data
> acqusition, first point is scaled with some factor, which is to
> eliminate ridges.
> After I predict some first points of time domain data with the linear
> precdiction method and replace the experimental data with them, which is
> right : I should scale the first point as above or I don't need to do
> such a thing after linear prediction? 
> 
> Happy New Year.
> 
> Pharmacy, Seoul National University
> 
> 


From owner-structural-nmr@net.bio.net Mon Jan 06 22:00:00 1997
Path: biosci!ihnp4.ucsd.edu!swrinde!cs.utexas.edu!news.sprintlink.net!news-peer.sprintlink.net!news.mindspring.com!usenet
From: Wave@mindspring.com (Pieter Ibelings)
Newsgroups: bionet.structural-nmr
Subject: PTS160 Generators Still Available 4Sale
Date: 7 Jan 1997 04:44:41 GMT
Organization: www.mindspring.com/~wave
Lines: 32
Message-ID: <5askbp$1qf@camel2.mindspring.com>
NNTP-Posting-Host: user-168-121-20-253.dialup.mindspring.com
Mime-Version: 1.0
Content-Type: Text/Plain; charset=US-ASCII
X-Newsreader: WinVN 0.99.8 (x86 32bit)

** RE. PTS160 Generators Still Available 4Sale.**

I still have a couple of PTS 160 Synthesizers for sale. I am asking $650 plus 
shipping for these real clean units. These are very clean with some minor 
scratches.I guarantee that they are in full working condition at time
of receipt. All the units have the TCXO, GPIB, BCD, Rack mount,
1Hz resolution and option X27??? These are surplus units from GE Medical 
Systems. All reasonable forms of payment are available (money orders or  
checks). If you want to send us a PO, we can send you the unit and then you 
can send us the check within a reasonable ammount of time. A friend of mine is 
taking care of the sales and he can be reached at the address and phone below. 
Please give him a call if you are ready to order and tell him that I refered 
you. If you require any additional information please do not hesitate to send 
me some e-mail.


Thanks

Pieter
wave@mindspring.com


********************************
Video Audio Repair Service Inc.
William E. Flake Jr. (Bill)
6658 B Hillandale Drv.
Lithonia, GA
30058

Phone (770) 482-2989
Tue-Fri 10:00am to 7:30pm


From owner-structural-nmr@net.bio.net Mon Jan 06 22:00:00 1997
Path: biosci!ihnp4.ucsd.edu!munnari.OZ.AU!news.ecn.uoknor.edu!feed1.news.erols.com!howland.erols.net!news.sprintlink.net!news-peer.sprintlink.net!news-peer.gsl.net!news.gsl.net!news-stkh.gsl.net!news.gsl.net!sn.no!nntp.uio.no!nntp-trd.UNINETT.no!daresbury!not-for-mail
From: Marc-Andre.Delsuc@cbs.univ-montp1.fr (Marc-Andre Delsuc)
Newsgroups: bionet.structural-nmr
Subject: Re: NMR to Matlab
Date: 7 Jan 1997 14:29:19 -0000
Lines: 35
Sender: lpddist@mserv1.dl.ac.uk
Distribution: bionet
Message-ID: <5atmjv$2um@mserv1.dl.ac.uk>
X-Sender: mad@tome.cbs.univ-montp1.fr
Original-To: str-nmr@dl.ac.uk

>Dear NMR-experts,
>can anybody help me with a program that converts 2D spectra (Bruker and
>Felix format) in
>Matlab-matrices and/or vice versa?
>Thanks a lot.
>Andr=E9
>
>------------------------------------------------------------------
>Andre Pampel
>University of Leipzig
>Faculty of Physics and Geosciences
>Linnestrasse 5
>04103 Leipzig
>Germany
>------------------------------------------------------------------

You may want to try the Gifa package. It is a complete NMR processing progra=
m
running on all kind of Unix machines (including Linux).
It permits to read Varian and Bruker files, and is able to read and write
Matlab matrices (ascii format).
check at :
http://www.cbs.univ-montp1.fr/GIFA/

Good luck !
Marc-Andr=E9

_________________________________________________________________________
Marc-Andre' Delsuc                        Centre de Biochimie Structurale
Marc-Andre.Delsuc@cbs.univ-montp1.fr                 Faculte de Pharmacie
tel : (33) (0) 467 04 34 36                       15 av, Charles Flahault
fax : (33) (0) 467 52 96 23                       34060 Montpellier cedex
www : http://www.cbs.univ-montp1.fr                                FRANCE



From owner-structural-nmr@net.bio.net Mon Jan 06 22:00:00 1997
Path: biosci!ihnp4.ucsd.edu!swrinde!howland.erols.net!feed1.news.erols.com!cwix!uunet!in1.uu.net!160.45.4.4!fu-berlin.de!news-ber1.dfn.de!news-lei1.dfn.de!news.uni-leipzig.de!news
From: André Pampel <anpa@rz.uni-leipzig.de>
Newsgroups: bionet.structural-nmr
Subject: NMR to Matlab
Date: Tue, 07 Jan 1997 08:42:23 -0800
Organization: Uni Leipzig
Lines: 15
Message-ID: <32D27CEF.F98@rz.uni-leipzig.de>
NNTP-Posting-Host: fun.exphysik.uni-leipzig.de
Mime-Version: 1.0
Content-Type: text/plain; charset=iso-8859-1
Content-Transfer-Encoding: 8bit
X-Mailer: Mozilla 3.01Gold (Win16; I)

Dear NMR-experts,
can anybody help me with a program that converts 2D spectra (Bruker and
Felix format) in
Matlab-matrices and/or vice versa?
Thanks a lot.
André

------------------------------------------------------------------
Andre Pampel
University of Leipzig
Faculty of Physics and Geosciences
Linnestrasse 5
04103 Leipzig
Germany
------------------------------------------------------------------

From owner-structural-nmr@net.bio.net Wed Jan 08 22:00:00 1997
Path: biosci!daresbury!lyra.csx.cam.ac.uk!uknet!usenet1.news.uk.psi.net!uknet!EU.net!news.bbnplanet.com!cpk-news-hub1.bbnplanet.com!portc02.blue.aol.com!audrey01.news.aol.com!not-for-mail
From: harrisonat@aol.com (HarrisonAT)
Newsgroups: bionet.structural-nmr
Subject: Nalorac ENC Symposium
Date: 9 Jan 1997 23:05:25 GMT
Organization: AOL http://www.aol.com
Lines: 20
Message-ID: <19970109230300.SAA22079@ladder01.news.aol.com>
NNTP-Posting-Host: ladder01.news.aol.com
X-Admin: news@aol.com

               ADVANCES IN NMR APPLICATIONS SYMPOSIUM


You are invited to attend the 5th Annual ADVANCES IN NMR APPLICATIONS
SYMPOSIUM featuring the latest developments in experimental techniques to
be held prior to ENC at the Omni Rosen Hotel, Ballroom D & E, 9840
International Drive, (located next to the Clarion Plaza Hotel, site of the
38th ENC) on Sunday, March 23, 2:30 to 6:00 p.m.

The agenda includes a presentation of recent results by leading NMR
experimentalists concerning applications of pulsed field gradient and
classical NMR techniques with both large and small molecular systems.  The
results obtained will be of interest to all liquid state NMR
Spectroscopists.

Request a detailed program or RSVP by contacting Chris Tierney, Nalorac s
ENC Coordinator. at NALORAC, 841-A  Arnold Drive, Martinez, CA 94553 
Phone:  (510) 229-3501 Fax: (510) 229-1651, Email: 
christierney@nalorac.com


From owner-structural-nmr@net.bio.net Wed Jan 08 22:00:00 1997
Path: biosci!agate!howland.erols.net!cam-news-hub1.bbnplanet.com!news.bbnplanet.com!cpk-news-hub1.bbnplanet.com!mindspring!news.mindspring.com!usenet
From: Wave@mindspring.com (Pieter Ibelings)
Newsgroups: bionet.structural-nmr
Subject: PTS160 Synthesizers for sale. See pictures @
Date: 9 Jan 1997 04:09:05 GMT
Organization: www.mindspring.com/~wave
Lines: 36
Message-ID: <5b1r11$9dk@camel5.mindspring.com>
NNTP-Posting-Host: user-168-121-20-253.dialup.mindspring.com
Mime-Version: 1.0
Content-Type: Text/Plain; charset=US-ASCII
X-Newsreader: WinVN 0.99.8 (x86 32bit)

** RE. PTS160 Generators Still Available 4Sale.**

Pictures available at:
http://www.mindspring.com/~wave/sale.html

I still have a couple of PTS 160 Synthesizers for sale. I am asking $650 plus 
shipping for these real clean units. These are very clean with some minor 
scratches.I guarantee that they are in full working condition at time
of receipt. All the units have the TCXO, GPIB, BCD, Rack mount,
1Hz resolution and option X27??? These are surplus units from GE Medical 
Systems. All reasonable forms of payment are available (money orders or  
checks). If you want to send us a PO, we can send you the unit and then you 
can send us the check within a reasonable ammount of time. A friend of mine is 
taking care of the sales and he can be reached at the address and phone below. 
Please give him a call if you are ready to order and tell him that I refered 
you. If you require any additional information please do not hesitate to send 
me some e-mail.


Thanks

Pieter
wave@mindspring.com


********************************
Video Audio Repair Service Inc.
William E. Flake Jr. (Bill)
6658 B Hillandale Drv.
Lithonia, GA
30058

Phone (770) 482-2989
Tue-Fri 10:00am to 7:30pm



From owner-structural-nmr@net.bio.net Sun Jan 12 22:00:00 1997
Path: biosci!Roche.COM!alfred.ross
From: alfred.ross@Roche.COM (Alfred Ross)
Newsgroups: bionet.structural-nmr
Subject: Deuteration of Glucose
Date: 13 Jan 1997 07:46:14 -0800
Organization: Roche Basel AG
Lines: 26
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <32DA5891.41C6@roche.com>
NNTP-Posting-Host: net.bio.net

Dear netters,


does anybody know a reference/recipe how to perform deuteration of
glucose.

Tanks a lot from

Alfred



-- 
	*************************************
	Dr. Alfred Ross
	NMR-Spectroscopist

	e-mail:	alfred.ross@roche.com	       		*******
	Phone:	CH-(0)61-6887029	       	       *       *
	Fax:	CH-(0)61-6887408	       	      *	 ROCHE  *
					      	       *       *		
	Mail:	F. Hoffmann-LaRoche AG	       		*******
		   (A. Ross - PRPI)
		        Postfach
		     CH-4070 Basel
	*************************************

From owner-structural-nmr@net.bio.net Sun Jan 12 22:00:00 1997
Path: biosci!SODIUM.BAS.ROCHE.COM!rossa
From: rossa@SODIUM.BAS.ROCHE.COM ("Alfred.Ross@Roche.com")
Newsgroups: bionet.structural-nmr
Subject: Deuteration of Glucose
Date: 13 Jan 1997 07:58:35 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 24
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <9701131647.ZM29386@sodium.bas.roche.com>
NNTP-Posting-Host: net.bio.net

Dear netters,

does anybody know a reference/recipe about the deuteration of glucose.

Thanks in advance


Alfred

-- 
	*************************************
	Dr. Alfred Ross
	NMR-Spectroscopist

	e-mail:	alfred.ross@roche.com	       		*******
	Phone:	CH-(0)61-6887029	       	       *       *
	Fax:	CH-(0)61-6887408	       	      *	 ROCHE  *
					      	       *       *		
	Mail:	F. Hoffmann-LaRoche AG	       		*******
		   (A. Ross - PRPS)
		        Postfach
		     CH-4070 Basel
	*************************************


From owner-structural-nmr@net.bio.net Sun Jan 12 22:00:00 1997
Path: biosci!biosci!not-for-mail
From: Bridget.Mabbutt@mq.edu.au
Newsgroups: bionet.structural-nmr,sci.techniques.mag-resonance
Subject: NMR postdoc position
Date: 12 Jan 1997 16:51:43 -0800
Organization: Macquarie University
Lines: 53
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <32D5B5B8.2BC0@mq.edu.au>
NNTP-Posting-Host: net.bio.net
Xref: biosci bionet.structural-nmr:1679 sci.techniques.mag-resonance:1926

The following position will shortly be advertised:
MUCAB Protein Structure Group
School of Chemistry
Macquarie University, Sydney, Australia

Protein NMR Spectroscopist

Salary:  Research Associate ($38,092 - 40,889, subject to approval) 


A two-year postdoctoral position is available immediately to determine 
the NMR structure of various proteins within the Protein Structure 
Group at Macquarie University, Sydney.  The successful applicant will 
primarily work to define a protein fold thought to be representative 
of a new family of cell-cell adhesion molecules.  Spectra obtained to 
date on this surface glycoprotein (MW=15kDa) are of very high quality 
(Zachara et al., Eur. J. Biochem. 238, 511-518, 1996) and milligram 
quantities of the 15N-labelled recombinant protein are routinely 
prepared in this laboratory.

The Protein Structure Group is focussed on several proteins involved 
in cellular adhesion and regulation.  It is staffed by a 
multidisciplinary team encompassing molecular biology, protein 
expression and NMR structure determination.  We are a component of the 
Macquarie University Centre for Analytical Biotechnology, and are 
closely associated with the Australian Proteome Analysis Facility, 
which together form the major centre for protein research in the 
Sydney region.  The Group is also part of IBiS (Initiative in 
Biomolecular Structure), a multidisciplinary forum addressing NMR, 
crystallography and computer techniques in structural biology.

A new 600 MHz NMR spectrometer will be installed in the School of 
Chemistry in 1997.  In the interim, generous access to a DMX-600 
spectrometer (with pulse-field gradient capability) is available 
nearby at the University of New South Wales.  Other resources include 
state-of-the art protein purification and analysis equipment,  and a 
cluster of dedicated Silicon Graphics workstations.  

Applicants should have a Ph.D. in Chemistry, Biochemistry or related 
field.  Experience in the recording and analysis of heteronuclear 
multidimensional NMR spectra of proteins is essential, as is 
familiarity with UNIX and NMR processing packages.

Appointment is available immediately, initially until December 1997, 
and is renewable subject to grant funding.  Applications will close 
early February 1997.   

For further information concerning applications contact:
Dr Bridget Mabbutt, School of Chemistry, 
email: bridget.mabbutt@mq.edu.au, 
telephone:  (61 2) 9850 8282, 
fax: (61 2) 9850 8313


From owner-structural-nmr@net.bio.net Tue Jan 14 22:00:00 1997
Path: biosci!daresbury!bioftp.unibas.ch!infobiogen.fr!jussieu.fr!u-psud.fr!usenet
From: Mark Blight <blight@igmors.u-psud.fr>
Newsgroups: bionet.structural-nmr
Subject: Help needed with Flame Spec
Date: Wed, 15 Jan 1997 19:39:29 +0100
Organization: Universite Paris-Sud, France.
Lines: 54
Message-ID: <32DD2461.786B@igmors.u-psud.fr>
NNTP-Posting-Host: bmec2.igmors.u-psud.fr
Mime-Version: 1.0
Content-Type: text/plain; charset=iso-8859-1
Content-Transfer-Encoding: quoted-printable
X-Mailer: Mozilla 2.0 (Macintosh; I; 68K)

Hello,

I am a molecular biologist by training and so am a bit rusty on some of the=
 apparatus and =

techniques that can be used for more chemical type studies.

However, I am interested in determining the "purity" of a complex mixture (=
a plant extract) =

with respect to other contaminating mixtures from, for example, other plant=
 species. I also =

need to process many samples with great rapidity and so organic extraction =
and HPLC etc =

are probably too slow. I was wondering if the absorption spectra (like for =
stellar objects) of =

a reference sample could be compared to an unknown and additional absorptio=
n bands =

subtracted to help identify the possible contaminant by comparison to a dat=
abank of =

potential contaminating substances (or mixtures). Like I said, I'm not very=
 familiar with =

the techniques that might be employed or the equipment. So, would I need a =
flame =

spectrometer (or is that a photometer?) to generate the absorption spectra?=
 And can anyone =

with more experience give me an idea if this might work?

Thanks for any suggestions,

Please reply by e-mail.

      Mark
_______________________________________
Dr. Mark A. Blight,
Institut de G=E9n=E9tique et Microbiologie,
CNRS URA 1354,
B=E2timent 409,
Universit=E9 de Paris XI,
91405 Orsay cedex,
France.

Tel:        +33 1 69 15 66 99
Fax:       + 33 1 69 15 78 08
e-mail:    BLIGHT@IGMORS.U-PSUD.FR
_______________________________________

From owner-structural-nmr@net.bio.net Thu Jan 16 22:00:00 1997
Path: biosci!CASPUR.IT!r.ragno
From: r.ragno@CASPUR.IT (Gianluca Sbardella)
Newsgroups: bionet.structural-nmr
Subject: Is OrgChem still "alive" ???
Date: 16 Jan 1997 23:41:39 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 26
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.OSF.3.95.970117082848.15988A-100000@naxos.caspur.it>
Reply-To: Gianluca Sbardella <r.ragno@caspur.it>
NNTP-Posting-Host: net.bio.net


Dear Netters,
I'm sending this question to a few different mail lists: my apologies for
any repetitiion!!!

The question is this:

Does anybody know if ORGCHEM (Organic Chemistry Mail List) IS STILL ALIVE?

Thank you all in advance,

Gianluca Sbardella

    ***************************************************************
    *                                                             *
    * Dr. Gianluca Sbardella     E-mail: r.ragno@caspur.it        *
    * Dip. Studi Farmaceutici                                     *
    * Universita' "La Sapienza"   Phone: 39-6-49913814            *
    * P.le A. Moro, 5               Fax: 39-6-491491              *
    * 00185 Roma                                                  *
    * ITALY                                                       *
    *                                                             *
    ***************************************************************




From owner-structural-nmr@net.bio.net Sun Jan 19 22:00:00 1997
Path: biosci!rutgers!gatech!csulb.edu!hammer.uoregon.edu!arclight.uoregon.edu!feed1.news.erols.com!howland.erols.net!news2.digex.net!news5.digex.net!news1.cstone.net!Skuzzy.cstone.net!not-for-mail
From: Dave Dolak <crystal@cstone.net>
Newsgroups: bionet.structural-nmr
Subject: NMR & aggregation
Date: Mon, 20 Jan 1997 10:55:21 -0500
Organization: Protein Solutions, Inc.
Lines: 12
Message-ID: <32E39569.17AD@cstone.net>
Reply-To: crystal@cstone.net
NNTP-Posting-Host: dialin92.cstone.net
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.0Gold (Win95; I)

How detrimental is a small amount of large aggregation during NMR
analysis?  I've heard some people say that it can be very detrimental
while others argue that it is not that important to eliminate.  Which is
it?  Can there be some aggregation that doesn't matter but above some
'threshold' it can interfere?

I'd like to better understand what limits there are on the presence of
aggregates for successful NMR analysis.
Thanks,
Dave Dolak
Protein Solutions, Inc.
70651.3674@compuserve.com

From owner-structural-nmr@net.bio.net Sun Jan 19 22:00:00 1997
Path: biosci!PHOENIX.PRINCETON.EDU!ipelczer
From: ipelczer@PHOENIX.PRINCETON.EDU (Istvan Pelczer)
Newsgroups: bionet.structural-nmr
Subject: Re: NMR & aggregation
Date: 20 Jan 1997 09:39:26 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 47
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SUN.3.91.970120122810.23463B-100000@phoenix.princeton.edu>
References: <32E39569.17AD@cstone.net>
NNTP-Posting-Host: net.bio.net


Dear Dave,

I am sure there will be people with more experience and more quantitative 
advise than what I can offer here, but let me put down my two cents...
I believe aggregation is most of a problem if it is uncontrolled.  The 
primary result is broadening lines, e.g., faster relaxation, enhanced 
spin diffusion, etc.  Dynamic characterization of the structure can give 
misleading information, too.  Symmetry of the aggregates can be a issue, 
as well.
I don't think there is a general "threshold" to be set;  should you have 
a chance, work with monomers at low enough concentration (as low as 0.2-0.5 
mM/L can be used with modern high-field instruments -- however, this is 
the edge of capabilities).
However, every sample is an individual, and practical compromises have to 
be considered.  A reasonable spectrum is better than no spectrum;  just 
be careful with the conclusions...
Good luck, and all the best,

Istvan

wwww,wwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwww
Istvan Pelczer, Ph.D.		       		Email: ipelczer@princeton.edu
Senior NMR Spectroscopist
Princeton University
Department of Chemistry, Frick Lab.,	 		 ph#  (609) 258 2342
Washington Road						fax#  (609) 258 6746
Princeton,  NJ 08544, USA



On Mon, 20 Jan 1997, Dave Dolak wrote:

> How detrimental is a small amount of large aggregation during NMR
> analysis?  I've heard some people say that it can be very detrimental
> while others argue that it is not that important to eliminate.  Which is
> it?  Can there be some aggregation that doesn't matter but above some
> 'threshold' it can interfere?
> 
> I'd like to better understand what limits there are on the presence of
> aggregates for successful NMR analysis.
> Thanks,
> Dave Dolak
> Protein Solutions, Inc.
> 70651.3674@compuserve.com
> 
> 

From owner-structural-nmr@net.bio.net Mon Jan 20 22:00:00 1997
Path: biosci!daresbury!lyra.csx.cam.ac.uk!news.ox.ac.uk!news
From: Isabelle Phan <phan@bioch.ox.ac.uk>
Newsgroups: bionet.structural-nmr
Subject: Re: NMR & aggregation
Date: Tue, 21 Jan 1997 12:06:08 +0000
Organization: Oxford Centre for Molecular Sciences
Lines: 20
Message-ID: <32E4B130.794BDF32@bioch.ox.ac.uk>
References: <32E39569.17AD@cstone.net> <Pine.SUN.3.91.970120122810.23463B-100000@phoenix.princeton.edu>
NNTP-Posting-Host: nmrh.ocms.ox.ac.uk
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.01 (X11; I; SunOS 4.1.3 sun4m)

Dear Dave,

As a follow up of Istvan Pelczer's post: Shurr et al., J.Magn.Reson.,
B105, 211-224, discuss the theoretical effect of aggregation on
relaxation measurements. They seem to say that 20% aggregation leads
already to significant errors when extracting dynamic parameters from
relaxation data.
On my own sample, I found that a fourfold dilution to about .5mM led to
a 30% drop in the apparent correlation time (see current issue of J.
Biomol. NMR).

Cheers,

Isabelle
-- 
____________________________________________________________________
   Dr. Isabelle Phan                       phan@bioch.ox.ac.uk
   Oxford Centre for Molecular Sciences    http://www.ocms.ox.ac.uk/ 
   University of Oxford                    voice: 44-1865-275773 
   OXFORD OX1 3QU, U.K.                    fax:   44-1865-275253

From owner-structural-nmr@net.bio.net Mon Jan 20 22:00:00 1997
Path: biosci!PINES.MED.UTORONTO.CA!vincent
From: vincent@PINES.MED.UTORONTO.CA (Sebastien Vincent)
Newsgroups: bionet.structural-nmr
Subject: Re: Aggregation and HSQC Linewidths
Date: 21 Jan 1997 14:45:55 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 33
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701212243.RAA11934@pines.med.utoronto.ca>
NNTP-Posting-Host: net.bio.net

> In practice, when working with a new protein, we usually do not know much
> about the aggregation parameters (unless we have done some fairly
> sophisticated sedimentation analysis).

Martin,

you can know a lot if you get a nitrogen-15 labelled sample in an NMR tube and record 15N T2 (see ) 
relaxation data. This gives, in 1 or 2 days, a way to obtain the overall correlation time, and this 
is part of an answer (Note: If you have time and want to be sure, you can also record 15N T1 and 
15N1H NOE mexperiments).
For a ref., see Farrow et al, Biochem. 33 (1994) 5984-6003.

In the case of aggregation, you can expect a certain correlation time for a monomer, and acertainly 
more for a monomer.

> what amide proton line widths should we be looking
> for to satisfy ourselves that standard triple resonance experiments will
> work and structure determination will be feasible for a new protein?

A rule of thumb is that if your overall correlation time is lower than 15ns, you're in good shape. 
If more, think deuteration or get another cDNA...

Sebastien


(::) (::) (::) (::) (::) (::) (::) (::) (::) (::) (::) (::) (::) (::)
Sebastien Vincent                     University of Toronto
vincent@redfield.med.utoronto.ca      Department of Medical Genetics,
Voice: ++(416) 978-0642               1, King's College Circle,
Fax:   ++(416) 978-6885               Toronto, Ontario, CANADA M5S 1A8
www: http://abragam.med.utoronto.ca/~vincent/

"Half the world population has never made a phone call." (M. Murphy, 12.96)

From owner-structural-nmr@net.bio.net Mon Jan 20 22:00:00 1997
Path: biosci!INDIANA.EDU!mastone
From: mastone@INDIANA.EDU (Martin J. Stone)
Newsgroups: bionet.structural-nmr
Subject: Aggregation and HSQC Linewidths
Date: 21 Jan 1997 12:40:21 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 45
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701212040.PAA01633@belize.ucs.indiana.edu>
NNTP-Posting-Host: net.bio.net

It seems to me that the effect aggregation has on an NMR spectrum will
depend on:
- the average size of the aggregated species
- the distribution of sizes of aggregated species
- the exchange rates between various aggregated and non-aggregated species
etc.

I would like to change the direction of the discussion slightly.

In practice, when working with a new protein, we usually do not know much
about the aggregation parameters (unless we have done some fairly
sophisticated sedimentation analysis).  Therefore, if we see lousy spectra,
we are left wondering whether they could potentially be improved by changes
in sample conditions that overcome aggregation.  In practice, we tend to
try many different variations of temperature, pH, salt conditions,
cosolvents, sample concentration, etc. until we find a spectrum that is
good enough, and then we assume we have overcome the aggregation effects in
the spectrum.  So my question is this:

        HOW GOOD IS GOOD ENOUGH??

...or more specifically, what amide proton line widths should we be looking
for to satisfy ourselves that standard triple resonance experiments will
work and structure determination will be feasible for a new protein?

I would appreciate some leading references (or some personal anecdotes)
that would help to address this question.


thanks

Martin


=======================================
Martin J. Stone,
Assistant Professor of Biochemistry,
Department of Chemistry,
Indiana University,
Bloomington, IN 47405.
Tel. (812)855-6779
Fax. (812)855-8300
email. mastone@indiana.edu
www:  http://pooh.chem.indiana.edu/


From owner-structural-nmr@net.bio.net Mon Jan 20 22:00:00 1997
Path: biosci!rutgers!gatech!csulb.edu!hammer.uoregon.edu!news-peer.gsl.net!news.gsl.net!howland.erols.net!news.bbnplanet.com!cam-news-hub1.bbnplanet.com!news.idt.net!news.cerf.net!ihnp4.ucsd.edu!scripps.edu!obelix!weber
From: weber@obelix.scripps (Christoph Weber)
Newsgroups: bionet.structural-nmr
Subject: Re: NMR & aggregation
Date: 21 Jan 1997 19:53:41 GMT
Organization: The Scripps Research Institute
Lines: 35
Message-ID: <5c36s5$ivd@riscsm.scripps.edu>
References: <32E39569.17AD@cstone.net>
Reply-To: weber@scripps.edu
NNTP-Posting-Host: obelix.scripps.edu
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
Keywords: protein


In article <32E39569.17AD@cstone.net>, Dave Dolak <crystal@cstone.net> writes:
> How detrimental is a small amount of large aggregation during NMR
> analysis?  I've heard some people say that it can be very detrimental
> while others argue that it is not that important to eliminate.  Which is
> it?  Can there be some aggregation that doesn't matter but above some
> 'threshold' it can interfere?

The problem is that 'aggregation' is a very broad term and in most
instances cannot be defined and/or measured in a sufficiently precise
way. If there are aggregates that show up in gel electrophoresis, then
they should and can be eliminated. 

However, in most cases we conclude that there is some 'aggregation' 
from circumstantial evidence, like overly broad lines in the NMR
spectrum. Since we don't even know what we're talking about really,
it is impossible to say at the outset how much is to much.

There are cases where we know what to expect. Camelized Ig VH domains
come to mind, for example, where there may be residual dimerization.
Having a precise idea of what 'aggregation' we are looking at can make
our job much easier.

As Istvan says, what really matters in the end is whether you can get 
NMR data of sufficiemt quality to do your analysis and whether you can 
circumvent additional problems, such as intermolecular NOEs.
Everything else is rather moot.

Christoph
|  Dr. Christoph Weber              Sen. Research Associate
|  Dept.of Molecular Biology, MB9   619-784-9869 (phone)
|  The Scripps Research Institute   619-784-2857 (FAX)
|  La Jolla  CA  92037-1027         weber@scripps.edu        
|  http://www.scripps.edu/~chazin/people/cw.html                   


From owner-structural-nmr@net.bio.net Mon Jan 20 22:00:00 1997
Path: biosci!MAILBOX.SYR.EDU!lpappala
From: lpappala@MAILBOX.SYR.EDU (Lucia  Pappalardo)
Newsgroups: bionet.structural-nmr
Subject: Bruker pulse sequences
Date: 21 Jan 1997 06:43:16 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 10
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SOL.3.95.970121091743.1084A-100000@forbin.syr.edu>
Reply-To: Lucia  Pappalardo <lpappala@mailbox.syr.edu>
NNTP-Posting-Host: net.bio.net


Dear Netters,
I am looking for the  pulse program of a 3D TOCSY/NOE w/ water suppression
using gradients suitable for an Avance Bruker DRX-500 w/
3-axis gradient probehead.   
Thanks in advance
Lucia




From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!MRIRIS.ROCKEFELLER.EDU!cowburn
From: cowburn@MRIRIS.ROCKEFELLER.EDU ("David Cowburn")
Newsgroups: bionet.structural-nmr
Subject: Aggregation phenomena, relaxation etc.
Date: 22 Jan 1997 06:23:40 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 33
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <9701220922.ZM5658@mriris.rockefeller.edu>
Reply-To: cowburn@rockvax.rockefeller.edu
NNTP-Posting-Host: net.bio.net

The ongoing discussion covers two slightly seperate issues concerned with the
same physical phenomena.  One issue is how to optimise sample conditions for
narrow linewidths, and at what point is deviation from the expected values a
problem.  Our own experience is that screening a range of conditions using
unlabelled or N15 labeled material is vital, and direct measurement of
aggregation state is critical.  Light scattering and low angle X-ray scattering
are difficult to do well and precisely; sedimentation equilibrium is very
valuable, but time consuming to do; amide hydrogen T2's and diffusion
measurments using gradients are empirically more useful to us.  As to what
conditions are unsuitable, if the target is of sufficient significance, then
you have to judge whether the critical experiments (e.g. CBCANH) can be done at
all, and if so, in what time.

The second issue is what are the effects of nonspecific aggregation on
relaxation and spectral properties, generally.  We have recently dealt with
such a system, and a paper will appear very shortly.  == Fushman, David;
Cahill, Sean M.; Cowburn, David (1997) "The main chain dynamics of the dynamin
pleckstrin homology (PH) domain in solution: Analysis of 15N relaxation with
monomer/dimer equilibration."  J. Mol. Biol.,266, 173-194. ==  The JMB issue
containing the paper should appear about Feb 1.

  While each protein system will have its own peculiarities, this paper deals
generally with the problem of nonspecific aggregation at the approx. 1mM
dissociation constant level, with relatively rapid exchange.  With the
relaxation equations modified for these conditions, the analysis can be done,
and the variations of tau(c) rationalised.  For this paper, sedimentation
equilibrium experiments were done, at NMR concentrations.


-- 
David Cowburn,	The Rockefeller University, 1230 York Avenue, NY, NY 10021, USA.
 	      Tel 212.327.8270. Fax 212.327.7566 http://mriris.rockefeller.edu/
		

From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!daresbury!not-for-mail
From: Paul Driscoll <driscoll@biochemistry.ucl.ac.uk>
Newsgroups: bionet.structural-nmr
Subject: aggregation or viscosity?
Date: 22 Jan 1997 11:57:50 -0000
Organization: Dept. Biochem. & Mol. Biol., UCL
Lines: 64
Sender: lpddist@mserv1.dl.ac.uk
Distribution: bionet
Message-ID: <5c4vbu$ahh@mserv1.dl.ac.uk>
Original-To: nmr@biochemistry.ucl.ac.uk, str-nmr@dl.ac.uk

I would like to put my oar in on this discussion about protein
aggregation. We have been working for some time now on protein systems
which demonstrate a concentration-dependent rotational correlation time
(based on 15N nuclear relaxation measurements). The obvious explanation
for this phenomenon is a fast exchange oligomerisation equilibria (e.g.
dimerisation or higher order association) that is enhanced at higher
protein concentration. I believe there are well documented examples of
this behaviour in the literature. I have been wondering if there are
other factors to be concerned about.

(Ignoring analytical ultracentrifugation for the moment...) 
In principle, estimates of molecular size or whether a protein is truly
monomeric in solution could be obtained by measurement of rotational
correlation times (tau_c) and comparison with calculations based on the
Stokes-Einstein equation or more sophisticated hydrodynamic models (e.g.
Beads, HYDRO, etc). At the end of the day one needs to know the
viscosity of the solvent to go into these calculations. My quandry is:
what defines the viscosity for this purpose in a heterogeneous
protein/solvent/buffer mixture? Is it safe to use the book value of the
viscosity for the solvent (H2O,D2O) or should one be concerned that at
least the buffer and no doubt the protein itself change in some manner
the structure of the solvent and alter the (macroscopic) viscosity? Or
is tau_c dependent more on local 'microscopic' protein-solvent
interactions?

I have thought a lot about this but can't decide if these 'viscosity'
factors are important. We tend to use deuterated TRIS/HCl as a buffer,
rather than phosphate. Also we often have 50mM-100mM salt present. Does
that mean we should invest time and money in measuring the viscosity of
the buffer explicitly to see if there are systematic effects on protein
tau_c values arising from this condition? Maybe. Anyone done anything
like this before? Perhaps the right conclusion is that the simple
hydrodynamic models are too simple - the solvation sphere of a protein
is likely a complex thing, different layers with graded characteristics,
and there could be specific salt effects of certain cations and anions
(maybe TRIS has some peculiarities in this regard?) which make a mockery
of these calculations. 

Against this view there are clearly proteins which behave as ideal
monomers, completely fitting the hydrodynamic models (e.g. ubiquitin),
which suggests that when one does not obtain sensible tau_c values for
other proteins then one either has (specific or non-specific)
aggregation, a 'floppy' or 'expanded' structure, an unusually high
degree of protein solvation, or there is something screwy with the
solvent viscosity. Perhaps others have some comments to make that would
help eliminate this last possibility.

Cheers,
Paul

P.S. A small plea. When people report tau_c values for proteins, I
strongly believe they should state the CALIBRATED temperature of the
sample in the spectrometer. Over recent times I have beome aware that
the reported probe temperature on the spectrometer console can be
significantly different from the actual sample temperature (perhaps one
should even take into account the RF heating effects of the pulse
sequences). I wonder whether the scatter one observes in a graph of
reported tau_c vs. molecular weight is exaggerated by this type of
discrepency.

---------------------------------------------------------------------
Paul Driscoll		    | Office (answer)phone: (44)-171 380 7035
Dept. Biochem. & Mol. Biol. |       Department fax: (44)-171 380 7193
University College London   |              driscoll@biochem.ucl.ac.uk

From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!internet!biosci!not-for-mail
From: biohelp (BIOSCI Administrator)
Newsgroups: bionet.structural-nmr
Subject: BIOSCI/bionet miniFAQ & Fundraiser
Date: 22 Jan 1997 02:00:14 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 239
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701221000.CAA27307@net.bio.net>
NNTP-Posting-Host: net.bio.net

(LAST REVISION: 30-JUL-95)

This BIOSCI "miniFAQ" is designed to answer the questions that come up
the *most frequently*.  The main BIOSCI FAQ (Frequently Asked
Questions) is accessible on the World Wide Web at URL
http://www.bio.net/.

If you can not find an answer to your question in this or other
documentation, the BIOSCI technical support staff answers e-mail
queries sent to

		       biosci-help@net.bio.net

We can only answer questions about the use of the newsgroups and
mailing lists.  We unfortunately do not have the staff to do Internet
information searches or answer scientific questions.  Please post
those to the appropriate BIOSCI/bionet newsgroups.


	Contents:
	--------
	0) BIOSCI NEEDS YOUR SUPPORT!!

	1) Using the WWW to access the BIOSCI/bionet newsgroups.

	2) What to do about "spams," i.e., junk mail, ads, etc.

	3) Examples of subscribing and unsubscribing to the mailing lists.

	4) The BIOSCI user address and research interest directory.


0) BIOSCI NEEDS YOUR SUPPORT!!
------------------------------
BIOSCI's government funding has been expended, and we are now
operating solely from advertising revenue that we have raised from our
Web site at http://www.bio.net/.  We need just a few minutes of your
time to help us serve you.

You can do two important things which will take very little time for
you individually and will immensely help us continue to help you.

First, please use our WWW system at http://www.bio.net/ to access the
archives.  You can post or reply to messages via your Web browser as
described in item #1 below.  Your usage helps attract sponsors. If you
contact any of our sponsors, please be sure to thank them for
supporting BIOSCI. It is critical for them to get this feedback if
they are to continue their sponsorship for the long term.

Second, if you work for a company or organization that provides
products or services of interest to the biology community, please pass
this message on to your marketing or marketing communications
department or other appropriate group.  Please ask them to help
support BIOSCI by sponsoring our Web site and explain the uses and
benefits of the system to the biology community. If they are
interested, they can then contact us for further information at our
tech support address, biosci-help@net.bio.net.


1) Using the WWW to access the BIOSCI/bionet newsgroups.
--------------------------------------------------------
As of 10 December 1995, all BIOSCI/bionet full newsgroups are
accessible through the World Wide Web (WWW) at URL http://www.bio.net.
One can read and reply publicly or privately to both recent postings
and archived messages through one's Web browser if it is configured
properly to send e-mail.  Each newsgroup is equipped with its own WAIS
index.  The main BIOSCI home page also has access to the BIO-JOURNALS
Table of Contents database WAIS index and the BIOSCI user address
database described in another item further below.


2) What to do about "spams," i.e., junk mail, ads, etc.
-------------------------------------------------------
BIOSCI is a set of parallel USENET newsgroups (the "bionet" groups),
mailing lists, and a hypermail archive at URL http://www.bio.net/.
The same postings are distributed on all media (except for a small
number of mailing-list-only groups at net.bio.net).  Unfortunately it
is becoming a despicable practice on the Internet (by a few people out
to make a fast buck) to do automated mass postings to thousands of
newsgroups and mailing lists.  These attempts to grab free advertising
are refered to as "spams" in the usual, somewhat boneheaded, net
terminology.  USENET is more susceptible to this practice, and many
spams originate on the USENET groups and then are passed on to the
mailing lists.  However, spammers also get lists of mailing addresses
and hit these too, so neither medium is immune.

What should you do personally if you get junk mail?
---------------------------------------------------
Just delete it and move on without reading it further.  Filing a
protest is becoming increasingly useless because spammers are often
disguising the addresses where the messages are sent from.  Unless you
really understand Internet mail systems, your attempt at protest by
sending replies to the message will often end up being sent to the
address of an innocent person that the spammer is victimizing.

What can BIOSCI/bionet do to protect its newsgroups?
----------------------------------------------------
The only solution currently available is to moderate the newsgroup.
If this newsgroup is already moderated, then you are in good shape.
Moderation protects the USENET distribution from about 95% of the
spams that are being sent to date and protects the mailing lists
completely.  Moderation means, however, that someone has to take the
time to review each message before it goes out.  We have set up
software here that simply allows the moderator to forward to an
address at net.bio.net messages that (s)he wishes to have distributed.
This takes no more time than that needed to read the message and pass
it on, say about 1 min. per message.

Most newsgroups currently have a discussion leader who is responsible
for their newsgroup.  The discussions leaders and their e-mail
addresses are listed in the BIOSCI Information Sheet which is
available on the Web at http://www.bio.net/.  If a newsgroup is being
hit with too many junk postings, please contact the discussion leader
for that group and see if there is interest in moderating the group.
Please do not assume that by simply posting a complaint to the
newsgroup itself, anyone on the BIOSCI staff will act on your
complaint.  With close to 100 newsgroups to run, the BIOSCI staff has
to rely on the discussion leaders of each newsgroup to report problems
directly to us at biosci-help@net.bio.net.

We will moderate any of our newsgroups if the discussion leader tells
us that the readership of the group wishes to do so and if a moderator
is willing to do the work.  For most BIOSCI/bionet groups, this
entails only a few minutes of work each day.

Moderating a newsgroup will resolve probably 95% of the junk postings
on the USENET distribution.  Unfortunately there are easy ways for
determined spammers to override the moderation mechanism on USENET,
but we can protect our e-mail subscribers from unwanted postings if
the newsgroup is moderated.  You can also access our newsgroups over
the WWW at URL http://www.bio.net.  While this Web interface will not
stop spammers from trying to post to the groups, this will give you
yet another way, besides using USENET news, to keep the junk out of
your personal mail files.  For those of you with local USENET news
systems, the Web interface will also give you faster access to new
newsgroups and recent postings.


3) Examples of subscribing and unsubscribing to the mailing lists.
------------------------------------------------------------------
PLEASE NOTE: The BIOSCI management does NOT act on
subscription/unsubscription requests that are posted improperly to the
newsgroups and mailing lists.  People who do this only bother everyone
on the lists to no avail.  Please be sure to follow the proper
procedures below.

Gory details are in the BIOSCI Information sheets on the Web at
http://www.bio.net.  Below we give an example utilizing the
METHODS-AND-REAGENTS list at both of our two BIOSCI sites:

Users in the Americas and Pacific Rim countries who use the BIOSCI
------------------------------------------------------------------
node at computer net.bio.net:
----------------------------

A) Determine the "listname" which is the <=8 character mail address
                                         ^^^^^^^^^^^^^
   for the group.  These can be found in the BIOSCI Info. Sheet.  For
   the METHODS-AND-REAGENTS group the mailing address is
   methods@net.bio.net.  The listname is the portion of the address to
   the left of the @ sign, i.e., "methods".  The listname is used with
   the "subscribe" and "unsubscribe" commands illustrated below.

B) Mail all commands in the body of a mail message addressed to
   biosci-server@net.bio.net.  Do NOT send commands to the newsgroup
   posting addresses!  Leave the Subject: line blank, any text on it
   will be ignored.

C) In the body of your message put one or more of the following
   commands with an "end" command on the last line, e.g.,

   subscribe methods
   unsubscribe methods
   end

   Do NOT put your e-mail address or other text on these lines.  The
   server only allows you to cancel your subscription if the address
   on your mail header matches the address on our mailing list.
   Please ask for help at biosci-help@net.bio.net if your address has
   changed, e.g., if you know you are on the list but the server tells
   you that you are not a member.


Users in Europe, Africa, and Central Asia who use the BIOSCI node at
--------------------------------------------------------------------
computer daresbury.ac.uk (also known as dl.ac.uk):
-------------------------------------------------

To subscribe and unsubscribe to/from the BIOSCI lists, you need to
specify the full USENET newsgroup name with "bionet-news." prepended.
The USENET newsgroup names are listed in the BIOSCI Information sheet
on the Web at http://www.bio.net/.  For the METHODS-AND-REAGENTS list
the USENET newsgroup name is bionet.molbio.methds-reagnts, thus the
appropriate commands are

    sub bionet-news.bionet.molbio.methds-reagnts

    unsub bionet-news.bionet.molbio.methds-reagnts

These commands are included in a message addressed to mxt@dl.ac.uk,
NOT to the newsgroup mailing addresses.  As usual, include the text in
the body of the message as text on the Subject: line is ignored.

To unsubscribe from all the lists at the UK node, use

    unsub bionet-news

Please note that if the address in the list is different than the one
in your mail message header, you will not be able to unsubscribe by
this method. If you have problems, please mail biosci@daresbury.ac.uk.


4) The BIOSCI user address and research interest directory.
-----------------------------------------------------------
Please take this opportunity to add your name, address, and research
interest information to the BIOSCI User Address Database if you have
not already done so.

You can fill out the address form directly through our Web page at URL
http://www.bio.net/adrform.html.

The address database is reindexed nightly for WWW access (the URL is
http://www.bio.net/).  If you are not directly on the Internet but can
reach it by e-mail, please use our waismail server to access the user
directory.  waismail use is described above.  You can also request a
user address form by e-mail from biosci-help@net.bio.net.

Please check your database entry from time-to-time to see if your
address information is still up-to-date.  Because of our limited
personnel resources, we ask that you resubmit a *complete* form to
revise your entry; we only replace complete entries and do not have
resources to edit old forms.

				Sincerely,

				Dave Kristofferson
				BIOSCI/bionet Manager

				biosci-help@net.bio.net

From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!biosci!not-for-mail
From: Reto Koradi <kor@mol.biol.ethz.ch>
Newsgroups: bionet.structural-nmr,bionet.software,bionet.molec-model,bionet.molbio.proteins,comp.graphics.visualization,comp.sys.sgi.apps
Subject: ANNOUNCE: MOLMOL 2.3
Date: 21 Jan 1997 21:38:56 -0800
Organization: Swiss Federal Institute of Technology (ETHZ)
Lines: 14
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <32E49735.2575@mol.biol.ethz.ch>
NNTP-Posting-Host: net.bio.net
Xref: biosci bionet.structural-nmr:1694 bionet.software:17661 bionet.molec-model:1333 bionet.molbio.proteins:9787 comp.graphics.visualization:9259 comp.sys.sgi.apps:14250

Release 2.3 of MOLMOL, a program for display and analysis of
macromolecular structures, is now available free of charge. Information
about the program and links to download sites can be found at:

  http://www.mol.biol.ethz.ch/wuthrich/software/molmol/

The program was developed as a joint project between BRUKER/Spectrospin
and the group of Prof. Wuthrich at the Institute for Molecular Biology
and Biophysics at the ETH Zurich. It runs on various UNIX workstations,
the ftp sites hold source code and binaries for Silicon Graphics, Sun,
IBM, Digital and Linux.
-- 
Reto Koradi (kor@mol.biol.ethz.ch, http://www.mol.biol.ethz.ch/~kor)


From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!daresbury!not-for-mail
From: Paul Driscoll <driscoll@biochemistry.ucl.ac.uk>
Newsgroups: bionet.structural-nmr
Subject: Re: Aggregation phenomena, relaxation etc.
Date: 22 Jan 1997 17:58:04 -0000
Organization: Dept. Biochem. & Mol. Biol., UCL
Lines: 32
Sender: lpddist@mserv1.dl.ac.uk
Distribution: bionet
Message-ID: <5c5kfc$8bm@mserv1.dl.ac.uk>
References: <97122145331.~INN-VSNa00197.bionet-news@dl.ac.uk>
Original-To: str-nmr@dl.ac.uk, David Cowburn <cowburn@mriris.rockefeller.edu>

Dear David,

Janet Thornton showed me a pre-print of your paper. It's excellent stuff, and we
enjoyed reading it. However, in your post you did not comment on what defines
viscosity in the 'Stokes-Einstein' sense. 
Do you take a view on this?

Cheers,
Paul

---------------------------------------------------------------------
Paul Driscoll		    | Office (answer)phone: (44)-171 380 7035
Dept. Biochem. & Mol. Biol. |       Department fax: (44)-171 380 7193
University College London   |              driscoll@biochem.ucl.ac.uk


> 
> The second issue is what are the effects of nonspecific aggregation on
> relaxation and spectral properties, generally.  We have recently dealt with
> such a system, and a paper will appear very shortly.  == Fushman, David;
> Cahill, Sean M.; Cowburn, David (1997) "The main chain dynamics of the dynamin
> pleckstrin homology (PH) domain in solution: Analysis of 15N relaxation with
> monomer/dimer equilibration."  J. Mol. Biol.,266, 173-194. ==  The JMB issue
> containing the paper should appear about Feb 1.
> 
>   While each protein system will have its own peculiarities, this paper deals
> generally with the problem of nonspecific aggregation at the approx. 1mM
> dissociation constant level, with relatively rapid exchange.  With the
> relaxation equations modified for these conditions, the analysis can be done,
> and the variations of tau(c) rationalised.  For this paper, sedimentation
> equilibrium experiments were done, at NMR concentrations.
>

From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!daresbury!bioftp.unibas.ch!ubaclu.unibas.ch!ubaclu.unibas.ch!nntp
Newsgroups: bionet.structural-nmr
Subject: aggregation /viscosity
Message-ID: <1997Jan22.181549.47295@yogi.urz.unibas.ch>
From: andrei <alexandrescu@ubaclu.unibas.ch>
Date: 22 Jan 97 18:15:49 MET
Organization: Biozentrum - U. Basel
Nntp-Posting-Host: skivacation.bioz.unibas.ch
X-Mailer: Mozilla 1.1N (Macintosh; I; PPC)
MIME-Version: 1.0
X-URL: news:bionet.structural-nmr
Content-Transfer-Encoding: 7bit
Content-Type: text/plain; charset=us-ascii
Lines: 22

Everyone in this discussion has focused on rotational diffusion (e.g. 
tau_c), which probably depends on billions of things.
  
An alternative way to measure viscosity/aggregation is to look at 
translational diffusion
see :
Alteri ... Byrd,
"Association of biomolecular systems via pulsed field gradient NMR
self-diffusion measurements"

Since this method uses 1D's I would imagine its a lot simpler than 3 
relaxation measurments and worries about R2ex and te.

P.S. Does anyone have experience with urea or other denaturants to 
reduce aggregation (i.e. enough urea to break up aggregates but not to 
unfold protein).  I know urea isn't very physiological but then neither 
are MPD, PEG, Ammonium Sulfate ...

Anyone know of any other methods to reduce aggregation (i.e. Chaps has 
been suggested)



From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!NMRSGI2.NCIFCRF.GOV!rabyrd
From: rabyrd@NMRSGI2.NCIFCRF.GOV (R. Andrew Byrd)
Newsgroups: bionet.structural-nmr
Subject: Aggregation and Diffusion
Date: 22 Jan 1997 08:21:40 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 86
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701221618.LAA24839@nmrsgi2.ncifcrf.gov>
NNTP-Posting-Host: net.bio.net


In light of recent postings to the newsgroup concerning
aggregation, and various discussions, we would like
to point out some elements of our recent work that we
used to address this problem.

Specifically, Martin Stone mentioned that unless one does 
some sophisticated sedimentation analyses, then the
aggregation state is unknown.  He also pointed out the
difficulties of knowing which direction to move in while
measuring lots of parameters, pH, temperature, salts, 
buffers, etc. and to know what the endpoint is.

Furthermore, the monitoring of parameters in an unlabeled
protein [...holding off on labels until you feel you
have the right conditions...] can be misleading if there
are things contributing to linewidth other than aggregation,
e.g. exchange with solvent or internal conformations, paramagnetic
metals contaminating the solvent.

Consequently, we felt it was important to monitor the bulk
behaviour of the particle under exactly the same conditions
as would be used for NMR experiments.  The approach is to
use translational diffusion measurements made in standard
pfg NMR probes.  We demonstrated that with even moderate
gradients [up to 30-40 G/cm] it is possible to measure
the diffusion constants of particles corresponding to 
molecular weights of 40-45 kDa.  For larger particles,
it is simple to see that they are greater than 45 kDa,
in which case you know the answer about using NMR!

The concepts behind the measurement of Ds are well established in
the literature.  Our approach was to adapt the work of
Gibbs and Johnson such that one can use current high-resolution
NMR probes and include one technique for water suppression.
There are various improvements and different ways to achieve
this, so I will only point you to the one approach which
we published:

	Altieri, Hinton, and Byrd; JACS 117:7566-7567 (1995)

A quantitative analysis of the Ds is not required, since 
that will depend on temperature, buffer viscosity, and possibly 
other factors. However, a comparative measurement can provide 
a clear answer to dimerization or higher oligimerization.  

The case of higher oligomerization causes a numerical instability 
in the analysis of the diffusion data (multi- component 
exponential fit where the difference in size is only
a factor of 2 or 3).  However, if high signal-to-noise ratios 
are obtained and one analyzes the residual plot [observed -
fit values], then it is possible to discern if the model
of a single species in solution fits reasonably.  The presence of 
systematic errors in the residual plot is an indication of 
non-compliance.  This characteristic should change when the 
conditions shift toward a single species, be it monomer, dimer, etc.

Nevertheless, as a 
RELATIVE measure, the Ds is still valid, and it is
reasonable to monitor the transition towards a monomeric
or stable dimeric state via the measured 'apparent' Ds.

One final point, I have stated that the analysis of 
exchange is complicated...this stems from the combination
of multiple exponential functions which differ in their
respective time constants by only, roughly, the order of
aggregation.

As pointed out by Paul Driscoll and others, this is a complex
issue.  I am less comfortable with comments which indicate
that as long as you can measure NOEs all is well, since
the effective buildup rate is dependent on the effective correlation
time. Any quantitative analysis of the NOE buildup rates must
also take these aggregation and exchange effects into account.

Finally, I would also like to point interested readers to the
work of Cynthia Larive's group and to Glen King's group (JBNMR
6:321-328 (1995)).

        Amanda Altieri   	Andy Byrd
        altieri@ncifcrf.gov	rabyrd@ncifcrf.gov
------------------
Dr. R. Andrew Byrd
Macromolecular NMR/MSL/ABL
TEL: 301-846-1407
FAX: 301-846-6195

From owner-structural-nmr@net.bio.net Tue Jan 21 22:00:00 1997
Path: biosci!MOLE.BIO.CAM.AC.UK!mjh2
From: mjh2@MOLE.BIO.CAM.AC.UK (Mark J Howard)
Newsgroups: bionet.structural-nmr
Subject: Re: aggregation or viscosity?
Date: 22 Jan 1997 06:51:59 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 70
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.95.970122145039.28349B-100000@mole.bio.cam.ac.uk>
NNTP-Posting-Host: net.bio.net



Just a few thoughts regarding Paul Driscoll's recent communication to
bionet.str-nmr:

> At the end of the day one needs to know the
> viscosity of the solvent to go into these calculations. My quandry is:
> what defines the viscosity for this purpose in a heterogeneous
> protein/solvent/buffer mixture? Is it safe to use the book value of the
> viscosity for the solvent (H2O,D2O) or should one be concerned that at
> least the buffer and no doubt the protein itself change in some manner
> the structure of the solvent and alter the (macroscopic) viscosity? Or
> is tau_c dependent more on local 'microscopic' protein-solvent
> interactions?

It is funny you should say this but I have encountered several instances
where text books and papers convince the reader that protein solutions
used in NMR are non-viscous due to the low sample concentration.  I have
seen several examples where this is not the case, and I am sure I am not
the only one.  These effects have been observed regardless of high or
low salt concentration.  With this in mind, if the protein concentrations
you are using are over 1mM, then there is every possibility that the
viscosity of H2O/D2O will be different from what you are actually dealing
with.  However, there can be some debate over the dependence of tau_c on
the macroscopic viscosity or 'local' viscosity.  I have heard of
good (as in narrow lines and fully resolable) protein spectra being
obtained from almost gel-like samples, this appears to back up a 'local'
rather than global viscosity.  

On the point of measuring viscosity, there are digital viscometers about,
but when I last looked into this in 1992 I was told they only measured
quite viscous mixtures and could not resolve viscosity differences much
below 10 times the viscosity of water but perhaps that has changed now.
The better method is to us an Oswald Viscometer (I remember my physical
chemistry practical well!).  These contraptions are very accurate, but you
must obtain the correct viscometer for the range of viscosity you intend
measuring.  You should still be able to get Oswald viscometers which
measure viscous differences of dilute solutions and water.  I think I may
know a possible source if you are interested.

> P.S. A small plea. When people report tau_c values for proteins, I
> strongly believe they should state the CALIBRATED temperature of the
> sample in the spectrometer. Over recent times I have beome aware that
> the reported probe temperature on the spectrometer console can be
> significantly different from the actual sample temperature (perhaps one
> should even take into account the RF heating effects of the pulse
> sequences). I wonder whether the scatter one observes in a graph of
> reported tau_c vs. molecular weight is exaggerated by this type of
> discrepency.

I could not agree more.  I personally feel that not enough experimental
details are given in many papers, especially important physical parameters
including the corrected temperature.

********************************************************
Dr. Mark Howard
Protein NMR Spectroscopy                    |
Dept. of Biochemistry                       | 
University of Cambridge                    || |
Tennis Court Road	      |		   || |    |
Cambridge, UK.		     ||	        | ||| |    |
CB2 1QW   		____|||||___/\__||||||||___|_|_ 

Tel: (01223) 333662  Fax: (01223) 333661
E-mail: mjh2@mole.bio.cam.ac.uk
********************************************************





From owner-structural-nmr@net.bio.net Wed Jan 22 22:00:00 1997
Path: biosci!rutgers!gatech!csulb.edu!hammer.uoregon.edu!news-xfer.netaxs.com!news.bbnplanet.com!cam-news-hub1.bbnplanet.com!howland.erols.net!usenet.kornet.nm.kr!newsfeed.dacom.co.kr!arclight.uoregon.edu!newsfeed.uk.ibm.net!news-m01.ny.us.ibm.net!newsfeed.de.ibm.net!02-newsfeed.univie.ac.at!03-newsfeed.univie.ac.at!fstgal00.tu-graz.ac.at!balu.kfunigraz.ac.at!balu!sengst
From: Helmut Sengstschmid <sengst@kfunigraz.ac.at>
Newsgroups: bionet.structural-nmr
Subject: control of pH
Date: Thu, 23 Jan 1997 17:01:47 +0100
Organization: Karl-Franzens-Universitaet Graz, Austria
Lines: 31
Message-ID: <Pine.SGI.3.93.970123165712.14842A-100000@balu>
NNTP-Posting-Host: balu.kfunigraz.ac.at
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; charset=US-ASCII
X-Sender: sengst@balu

Good evening!

I have not been in the field of peptide/protein NMR for long and so I lack
a lot of practical experience.
Recently I had the problem of measuring the pH of a protein solution.
What is the best way to do that? Do you use these small electrodes to fit
into the NMR tube?
Furthermore if you want to measure a spectrum at a given pH (e.g. pH 5.0)
what is the best way to get there? Do you use buffers (if yes, which) or
do you just add a few drops of HCl? 
Furthermore this whole issue gets complicated if you add some other
compounds like SDS (for simulating membranes) that may change the pH quite
severely and that cannot be used in combination with certain buffers (e.g.
SDS precipitates with phosphate).

Thanks for your replies,

Helmut Sengstschmid

-----------------------------
DI Helmut Sengstschmid
Institut fuer Organische Chemie
Karl Franzens Universitaet 
Heinrichstr. 28
A-8010 Graz
Austria

Tel.:   ++43 / 316 / 380 53 44
Fax:    ++43 / 316 / 38 33 40
E-Mail: sengst@kfunigraz.ac.at


From owner-structural-nmr@net.bio.net Wed Jan 22 22:00:00 1997
Path: biosci!UGA.CC.UGA.EDU!JLEE
From: JLEE@UGA.CC.UGA.EDU
Newsgroups: bionet.structural-nmr
Subject: protein solution viscosity
Date: 23 Jan 1997 06:55:40 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 11
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <970123.095503.EST.JLEE@UGA.CC.UGA.EDU>
NNTP-Posting-Host: net.bio.net

The question of what solution (solvent) viscosity to use is adequately
discussed in H.K.Schackmann's book "Ultracentrifugation in
Biochemistry". In this field the solvent viscosity is taken since the
macromolecules seldom encounter one another.

John Lee
Department of Biochemistry
University of Georgia
Athens GA 30602-7229 USA
Phone 706-542-1764
FAX   706-542-1738

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!BIBA.MED.TUFTS.EDU!sudha
From: sudha@BIBA.MED.TUFTS.EDU (Sudha Veeraraghavan)
Newsgroups: bionet.structural-nmr
Subject: handling proteins
Date: 23 Jan 1997 20:15:40 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 22
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701240411.XAA21042@biba.med.tufts.edu>
NNTP-Posting-Host: net.bio.net

Helmut,

The first thing you would want to do is to get the book edited by G.C.K. Roberts called "NMR of Macromolecules - A Practical Approach" (IRL Press, 1993).  In this book, the second chapter, written by W. Primrose, details painstakingly techniques and methodologies in handling protein samples for NMR.  

To directly answer your questions:  Yes, you need to have a buffer that is suitable for your purposes and one that will has a pKa in the range of pH you wish to keep your sample at.  Yes, you will need to use a pH electrode thin enough to insert into the NMR tube - so you can measure the pH in the NMR tube.  You may also pH the sample before transferring it into the NMR tube.  Adjusting the pH is a bit tricky.  One has to be very careful doing this.  In my experience with pHing proteins, HCl and NaOH are two reagents best not used.  Using a dual buffer system, e.g., phosphate-citrate or phosphate-acetate, reduces loss of sample to precipitation.   

Good luck
--Sudha
_______________________________________________________________________________

*******************************************************************************
!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!

Sudha Veeraraghavan, Ph.D.                 Phone: (617) 636-6873
Department of Biochemistry MV605           Fax  : (617) 636-6409
Tufts Univ. Sch. Medicine
136 Harrison Ave.                          E-mail:  sudha@biba.med.tufts.edu
Boston, MA 02111                                    sveerara@opal.tufts.edu

!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
*******************************************************************************
_______________________________________________________________________________

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Newsgroups: bionet.structural-nmr
Path: biosci!rutgers!utcsri!utnut!utinfo!bloch.med.utoronto.ca!not-for-mail
From: gardner@bloch.med.utoronto.ca (Kevin Gardner)
Subject: Re: aggregation /viscosity
X-Nntp-Posting-Host: bloch.med.utoronto.ca
Message-ID: <E4Gto9.24o@utcc.utoronto.ca>
Lines: 42
Sender: nntp@utcc.utoronto.ca (News)
Organization: UTCC Campus Access
X-Newsreader: TIN [UNIX 1.3 950824BETA PL0]
References: <1997Jan22.181549.47295@yogi.urz.unibas.ch>
Date: Thu, 23 Jan 1997 14:31:21 GMT

andrei (alexandrescu@ubaclu.unibas.ch) wrote:
: P.S. Does anyone have experience with urea or other denaturants to 
: reduce aggregation (i.e. enough urea to break up aggregates but not to 
: unfold protein).  I know urea isn't very physiological but then neither 
: are MPD, PEG, Ammonium Sulfate ...
: 
: Anyone know of any other methods to reduce aggregation (i.e. Chaps has 
: been suggested)

Aside from the facetious suggestion of raising temperature :), I've
got two others from a couple of papers:

1.  trifluoroethanol (TFE), used up to 15% v/v to break apart a troponin
C dimer:

"Calcium-induced dimerization of troponin C: mode of interaction and
use of TFE as a denaturant of quaternary structure"  Slupsky, C.M.
et al Biochemistry 34(1995): 7365-7375.

2.  partial proteolysis to remove non-specifically aggregating
segments --- this was demonstrated nicely in a figure
from Mike Summers' group's determination of the HIV capsid protein
structure (I believe Gitti, R.K. et al.  Science 273(1996): 231-235?).
By adding a small amt of protease to a protein solution and recording
1D 1H spectra over time a marked improvement was observed.



One last note in defense of small-angle X-ray scattering: I've found this
to be a wonderfully complementary approach to NMR as one can/should use
a protein sample at the same concentrations for these techniques.  I agree
that there is a problem with the relatively limited number of sites that
with scattering setups --- 

Cheers,
Kevin

-- 
*************************************************************************
Kevin Gardner                               gardner@bloch.med.utoronto.ca
University of Toronto             http://abragam.med.utoronto.ca/~gardner
Dept. of Medical Genetics & Microbiology   phone: 416-978-0642/FAX: -6885

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!daresbury!bioftp.unibas.ch!ubaclu.unibas.ch!ubaclu.unibas.ch!nntp
Newsgroups: bionet.structural-nmr
Subject: Re: control of pH
Message-ID: <1997Jan24.130323.47303@yogi.urz.unibas.ch>
From: Andrei Alexandrescu <alexandrescu@ubaclu.unibas.ch>
Date: 24 Jan 97 13:03:22 MET
References: <Pine.SGI.3.93.970123165712.14842A-100000@balu>
Organization: Biozentrum, U-Basel
Nntp-Posting-Host: secsy.bioz.unibas.ch
X-Mailer: Mozilla 1.1N (Macintosh; I; PPC)
MIME-Version: 1.0
X-URL: news:Pine.SGI.3.93.970123165712.14842A-100000@balu
Content-Transfer-Encoding: 7bit
Content-Type: text/plain; charset=us-ascii
Lines: 32

(1) The G.C.K. Roberts book "NMR of Macromolecules" is very good, and 
has a good introduction to sample preparation.

(2) This is probably heresy (please no e-mail) but personally, I'm not a 
big fan of buffers.  I think protein solution at 1-2 mM concentrations 
are usually quite good buffers already.  Unless you have a 1-sample NMR 
project you also have to consider what to do with the buffer if you want 
to concentrate the sample, change the pH, add Ca2+, etc...  Buffers also 
have some detrimental NMR effects (described in Roberts book). I will 
grant, however, that some proteins do behave pretty badly without 
buffers, so I guess it depends on the protein.  The biologists who gave 
you the protein will probably have suggestions (e.g. PBS).  Unless I was 
given the world's only 4 mg sample of quadruple labeled preciouslin that 
took 5 years to purify, however, I would be inclined to collect a 
spectrum of the protein without added buffers.

(3) While using electrodes that fit in an NMR tube is an option, its 
not the only one!  In MY HANDS these rather expensive electrodes seem to 
have a high proclivity to snap into little pieces. You also have to 
consider how to mix the sample in the NMR tube.  I prefer electrodes 
that fit in an Eppendorf tube.  You can pH the sample in the Eppendorf, 
and transfer it to the NMR tube.  You have to be careful, however, with 
"newly cleaned" NMR tubes since they may have been cleaned with 
something acidic.  Its a good idea to take the sample back out of the 
NMR tube and check its pH again after the initial transfer.

For changing pH, with HCl or DCl, dilution helps (e.g. dilute to 1M or 
less).  For really small pH changes you can touch the tip of your 
pippete to a kimwipe or other absorbant tissue to decrease the size of 
the drop (assuming your kimwipes are not a major source of impurities).



From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!SNEEZY.FHIS.NET!barry
From: barry@SNEEZY.FHIS.NET ("Barry Schweitzer")
Newsgroups: bionet.structural-nmr
Subject: Re: aggregation /viscosity
Date: 24 Jan 1997 06:05:29 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 36
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <9701240903.ZM4981@sneezy.fhis.net>
NNTP-Posting-Host: net.bio.net

On Jan 23,  2:31pm, Kevin Gardner wrote:
> Subject: Re: aggregation /viscosity
> andrei (alexandrescu@ubaclu.unibas.ch) wrote:
> : P.S. Does anyone have experience with urea or other denaturants to
> : reduce aggregation (i.e. enough urea to break up aggregates but not to
> : unfold protein).  I know urea isn't very physiological but then neither
> : are MPD, PEG, Ammonium Sulfate ...
> :
> : Anyone know of any other methods to reduce aggregation (i.e. Chaps has
> : been suggested)
>

You might want to check out the paper from Jim Baleja's group (Biochemistry
(1995) 34, 12126-12137).  They use a cocktail of urea and guanidinium
to keep a Ca2+ bound gla domain (otherwise known as the Rock of Gilbralter)
soluble.  If you're lucky, your protein will retain its structure in this
solution (which is definitely not physiological!), but lose its tendency to
aggregate.  It is somewhat problematic to knock down the large urea and
guanidine peaks, but we have found that it can be done using either a WET
sequence or excitation sculpting.

Good luck!

Barry

-- 
Barry Schweitzer, Ph.D.
Director 
Division of Molecular & Structural Biology 
Walt Disney Memorial Cancer Institute 
 at Florida Hospital
12722 Research Parkway
Orlando, FL 32826
Phone:	(407) 380-9977
FAX:	(407) 380-9978
email:	barry@sneezy.fhis.net

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!daresbury!not-for-mail
From: Paul Driscoll <driscoll@biochemistry.ucl.ac.uk>
Newsgroups: bionet.structural-nmr
Subject: Re: control of pH
Date: 24 Jan 1997 15:10:48 -0000
Organization: Dept. Biochem. & Mol. Biol., UCL
Lines: 32
Sender: lpddist@mserv1.dl.ac.uk
Distribution: bionet
Message-ID: <5cajdo$cgn@mserv1.dl.ac.uk>
References: <97124123520.~INN-QBPa00197.bionet-news@dl.ac.uk>
Original-To: str-nmr@dl.ac.uk, Andrei Alexandrescu <alexandrescu@ubaclu.unibas.ch>

Andrei Alexandrescu wrote:

> (2) This is probably heresy (please no e-mail) but personally, I'm not a
> big fan of buffers.  I think protein solution at 1-2 mM concentrations
> are usually quite good buffers already.  Unless you have a 1-sample NMR
> project you also have to consider what to do with the buffer if you want
> to concentrate the sample, change the pH, add Ca2+, etc...  Buffers also
> have some detrimental NMR effects (described in Roberts book). I will
> grant, however, that some proteins do behave pretty badly without
> buffers, so I guess it depends on the protein.  The biologists who gave
> you the protein will probably have suggestions (e.g. PBS).  Unless I was
> given the world's only 4 mg sample of quadruple labeled preciouslin that
> took 5 years to purify, however, I would be inclined to collect a
> spectrum of the protein without added buffers.
> 

Surely this is wrong. Proteins are only buffers themselves in that 
they adsorb or give up protons on acidic and basic residue sidechains. 
The consequences of this are that for a sample coming to equilibrium 
during an experiment (e.g. by slow and small amounts of precipitation) 
or titration of a ligand the pH changes slightly, and the chemical shifts 
of some side chain and even backbone NH signals are thereby likely to 
change. This can cause severe problems in further analysis of the 
spectra. I would always recommend the use of a 10-20mM exogeneous 
buffer if possible. Believe me, it can eliminate a lot of hassle 
later on.

Paul
---------------------------------------------------------------------
Paul Driscoll		    | Office (answer)phone: (44)-171 380 7035
Dept. Biochem. & Mol. Biol. |       Department fax: (44)-171 380 7193
University College London   |              driscoll@biochem.ucl.ac.uk

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!sfu.ca!dnaugler
From: dnaugler@sfu.ca (David Naugler)
Newsgroups: bionet.structural-nmr
Subject: Re: control of pH
Date: 24 Jan 1997 08:07:52 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 33
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701241608.QAA03555@beaufort.sfu.ca>
References: <1997Jan24.130323.47303@yogi.urz.unibas.ch>
NNTP-Posting-Host: net.bio.net

From Andrei Alexandrescu

> (2) This is probably heresy (please no e-mail) but personally, I'm not a 
> big fan of buffers.  I think protein solution at 1-2 mM concentrations 
> are usually quite good buffers already.  Unless you have a 1-sample NMR 
> project you also have to consider what to do with the buffer if you want 
> to concentrate the sample, change the pH, add Ca2+, etc...  Buffers also 
> have some detrimental NMR effects (described in Roberts book). I will 
> grant, however, that some proteins do behave pretty badly without 
> buffers, so I guess it depends on the protein.  The biologists who gave 
> you the protein will probably have suggestions (e.g. PBS).  Unless I was 
> given the world's only 4 mg sample of quadruple labeled preciouslin that 
> took 5 years to purify, however, I would be inclined to collect a 
> spectrum of the protein without added buffers.

Although this topic is beyond my current level of expertise, I am 
inclined to accept this point of view based on my own experience. In work 
with an intractable peptide, a fragment of an integral membrane protein, 
I found that I could prepare an acceptable NMR sample by repeated 
dialysis against distilled water. The peptide has a tendency to aggregate 
above a pH 4.1. The lyophilized peptide can then be prepared as an NMR 
sample at pH 6.49 with SDS.

In the case of a soluble protein rather than a protein subunit, SDS would 
not be necessary. I would think that a proper NMR sample could be 
prepared by the dialysis of any buffer compounds. Because of the presence 
of basic and acidic residues such a prepation should be self buffering 
and have a stable pH at least stable enough for good NMR.

David Naugler
Institute of Molecular Biology and Biochemistry
Simon Fraser University


From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!A.CHEM.UPENN.EDU!eshapiro
From: eshapiro@A.CHEM.UPENN.EDU ("Erik M. Shapiro")
Newsgroups: bionet.structural-nmr
Subject: Subscribe
Date: 24 Jan 1997 13:33:30 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 2
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.91.970124163322.17109A@a.chem.upenn.edu>
NNTP-Posting-Host: net.bio.net


Subscribe.

From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!MSI.COM!mjf
From: mjf@MSI.COM (Mark J Forster)
Newsgroups: bionet.structural-nmr
Subject: unsubscribe
Date: 24 Jan 1997 13:46:44 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 3
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701242146.NAA26873@iris70.msi.com>
NNTP-Posting-Host: net.bio.net


unsubscribe


From owner-structural-nmr@net.bio.net Thu Jan 23 22:00:00 1997
Path: biosci!charnwood.gb.astra.com!"/I=JM/G=Mark/S=Dixon/OU=Physical and Metabolic Sciences/"
From: "/I=JM/G=Mark/S=Dixon/OU=Physical and Metabolic Sciences/"@charnwood.gb.astra.com (Mark Dixon, Tel +44 1509 644 062)
Newsgroups: bionet.structural-nmr
Subject: 4D in H2O
Date: 24 Jan 1997 11:03:47 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 24
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <"91546142107991.41123.MRX*"@MHS>
NNTP-Posting-Host: net.bio.net


               Dear Protein Spectroscopists,
               
               I have searched the literature for variants of the 4D 13C/13C 
               HMQC-NOESY-HMQC with gradients, and I have hit upon a bit of 
               a problem.
               
               All of the sequences I have found operate in a D2O solution, 
               whereas I have _one_ sample in 90% H2O/10% D2O, which is not 
               amenable to solvent exchange.
               
               Can anyone tell me why it is not possible to run this 
               experiment in predominantly H2O ?  I am prepared to consider 
               the various tricks for water suppression, but I cannot see why 
               the water should be a problem.
               
               Thank you for your consideration,
               
               
               J. Mark Dixon
               ASTRA CHARNWOOD
               
               mark.j.m.dixon@charnwood.gb.astra.com
               

From owner-structural-nmr@net.bio.net Fri Jan 24 22:00:00 1997
Path: biosci!CARIBE.CHEM.UKY.EDU!cammers
From: cammers@CARIBE.CHEM.UKY.EDU (Arthur Cammers)
Newsgroups: bionet.structural-nmr
Subject: (none)
Date: 24 Jan 1997 20:08:37 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 63
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.93.970124231044.2721A-100000@caribe.chem.uky.edu>
NNTP-Posting-Host: net.bio.net

	I would like to invite all who are close enough for a day trip to
the University of Kentucky's Symposium on chemistry and molecular biology;
the topic this year is "The Role of Combinatorial Chemistry in Drug
Design."  Attendence is free of charge.
	We will have a series of four seminars on april 8th starting at
8:30 a.m.-1:00 p.m.  Check out the webpage.
http://www.chem.uky.edu/seminars/naff97.html
For those of you who are not www capable below are the speakers and short
point-wise outlines of their talks.
------------------------------------------------- 
-Arno Spatola (University of Louisville)

Title:  "Combinatorial Chemistry: The End of Rational Drug
Design? Lessons from Macrocyclic Libraries"

- basic combinatorial methods
- a Brief Historical Overview
- a "Realistic" View of the Methods-not a Panacea
- Why so much Industrial Attention?
- Parallel synthesis vs. mixtures
- The deconvolution problem
- Combinatorial Chemistry at the University of Louisville

-----                   -------
-Kim Janda (Scripps Institute and CombiChem)

Title: "From Combinatorics to Soluble Polymer Supported Synthesis"

- Encoded Combinatorial Chemistry
- Recursive Deconvolution
- Liquid Phase Combinatorial Synthesis of peptides,
- Peptidomimetics, and Small Molecules
- Multipolymeric Reactions
- Combizymes
- Combinatorial Polymer Synthesis
-----                   -------

-Jon Ellman (University of California at Berkeley)

Title: "Design, Synthesis, and Evaluation of Small Molecule Libraries"

- General Philosophy for the Design of Combinatorial Schemes
- Prudent Selection of Classes of Compounds for Combinatorial Approaches.
- Heuristic Bioactivity of Known Drugs as Starting Points in Combinatorial
approaches.
- Synthetic Approaches to Maximize Molecular Diversity

-----                   -------
-Dr. Mario Geysen (Glaxo Wellcome Pharmaceuticals)
- Dr. Geysen is credited as inventor of combinatorial synthesis.

Title: "Combinatorial Chemistry:  A New Paradigm for Drug Discovery."

- Combinatorial Chemistry at Glaxo Wellcome.
-------------------------------------------------------------------------

Arthur Cammers-Goodwin				
Assistant Professor of Chemistry		
Department of Chemistry-Organic Division 	
University of Kentucky				
acgood1@pop.uky.edu
http://www.chem.uky.edu/research/cammers/cammerscv.html


From owner-structural-nmr@net.bio.net Sat Jan 25 22:00:00 1997
Path: biosci!MAILBOX.UQ.OZ.AU!L.Miranda
From: L.Miranda@MAILBOX.UQ.OZ.AU (Leslie Miranda)
Newsgroups: bionet.structural-nmr
Subject: (none)
Date: 25 Jan 1997 21:52:11 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 4
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.OSF.3.95.970126155227.8261C-100000@dingo.cc.uq.oz.au>
NNTP-Posting-Host: net.bio.net

subscribe

Les Miranda


From owner-structural-nmr@net.bio.net Sun Jan 26 22:00:00 1997
Path: biosci!Merck.Com!yves_aubin
From: yves_aubin@Merck.Com (Yves Aubin)
Newsgroups: bionet.structural-nmr
Subject: Shielding
Date: 27 Jan 1997 14:50:22 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 52
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701272104.QAA16919@igw2>
NNTP-Posting-Host: net.bio.net

Dear netters,

Several months ago, there was an inquiry from Istvan about shielding monitors
from the stray fields of NMR magnets.   In the solution I proposed (a homebuilt
shield made of plywood and M6 shielding material) I mentionned the name of
the company from where we got the material but not the person who helped
us.  I guess I wanted to shield this person from too many phone calls...
In fact,
I apologize because I should have acknowledged his help in choosing the right
material for our needs.

This person is named Vahe Ohanian.  Vahe is an engineer at BCL Magnetics and
was very helpful in helping me getting a good shielding material for an
affordable
price (compare to NUmetal).  Talking with him recently,  he was telling me
that he
is testing a new software that will enable him to model magnetic fields and
allow
him to do better suggestions regarding shielding design and shielding material.
He told me that he would be keen in helping people in those issues.  Is
coordinates
are:

			Vahe Ohanian
			BCL-Magnetics
			5045 North Service Rd
			Burlington, ON, Canada,
			L7L 5H6
			phone: (905) 335-2530
			FAX:    (905) 335-8474
			email: magnetic@world.com

I thought that it would be useful to the NMR community to know someone who
knows
about magnetic shielding.

		Hope this helps!


			Yves


Yves Aubin Ph.D.
Merck Frosst Canada Inc.
P.O. Box1005
Pointe-Claire--Dorval
QC, H9R 4P8
tel: (514) 428-3931
fax: (514) 428-8615
email: yves_aubin@merck.com



From owner-structural-nmr@net.bio.net Sun Jan 26 22:00:00 1997
Path: biosci!rutgers!news.sgi.com!news.bbnplanet.com!su-news-hub1.bbnplanet.com!news.sprintlink.net!news-peer.sprintlink.net!howland.erols.net!surfnet.nl!highway.leidenuniv.nl!usenet
From: Aart Spilt <aart@lkeb.medfac.leidenuniv.nl>
Newsgroups: bionet.structural-nmr,sci.materials,sci.med,sci.med.physics,sci.med.radiology
Subject: Monel
Date: Mon, 27 Jan 1997 11:23:12 +0100
Organization: Leiden University, The Netherlands
Lines: 9
Message-ID: <32EC8210.2388@lkeb.medfac.leidenuniv.nl>
NNTP-Posting-Host: 913azl63.azl.nl
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.01 (Win95; I; 16bit)
Xref: biosci bionet.structural-nmr:1718 sci.materials:21123 sci.med:148763 sci.med.physics:5110 sci.med.radiology:8283

Hello,

I'm trying to find out anything about the magnetic characteristics of
Monel. It is an Alloy of Ni, C, Mn Fe, S, Si and Cu. Please give me all
the information you have. 

I will be very thankful with every kind of information.

Aart Spilt

From owner-structural-nmr@net.bio.net Sun Jan 26 22:00:00 1997
Path: biosci!SHIELDING.USI.UTAH.EDU!facelli
From: facelli@SHIELDING.USI.UTAH.EDU (Julio Facelli)
Newsgroups: bionet.structural-nmr
Subject: unsubscribe
Date: 27 Jan 1997 08:03:49 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 15
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701271602.JAA03629@shielding.usi.utah.edu>
NNTP-Posting-Host: net.bio.net

unsubscribe
________________________________________________________________________________

Julio C. Facelli, PhD.
Director 

Center for High Performance Computing (CHPC) 
66 SSB, The University of Utah
Salt Lake City, Utah 84112, USA. 

Phone    (801)-581-7529
Fax      (801)-585-5366
e-mail   facelli@shielding.usi.utah.edu

________________________________________________________________________________

From owner-structural-nmr@net.bio.net Mon Jan 27 22:00:00 1997
Path: biosci!rutgers!gatech!www.nntp.primenet.com!nntp.primenet.com!su-news-hub1.bbnplanet.com!news.bbnplanet.com!cpk-news-hub1.bbnplanet.com!portc02.blue.aol.com!newstf02.news.aol.com!audrey01.news.aol.com!not-for-mail
From: harrisonat@aol.com (HarrisonAT)
Newsgroups: bionet.structural-nmr
Subject: Nalorac Symposium
Date: 28 Jan 1997 22:49:05 GMT
Organization: AOL http://www.aol.com
Lines: 55
Message-ID: <19970128224900.RAA26848@ladder01.news.aol.com>
NNTP-Posting-Host: ladder01.news.aol.com
X-Admin: news@aol.com

ADVANCES IN NMR APPLICATIONS SYMPOSIUM - AGENDA

You are invited to attend the 5th Annual ADVANCES IN NMR APPLICATIONS
SYMPOSIUM featuring the latest developments in experimental techniques to
be held prior to ENC at the Omni Rosen Hotel, Ballroom D & E, 9840
International Drive, (located next to the Clarion Plaza Hotel, site of the
38th ENC) on Sunday, March 23, 1:15 to 6:00 p.m.

The Role of NMR in the Study of Drug Metabolism.
John Shockcor, John Lindon and Jeremy Nicholson, Glaxo Wellcome

Carbon 13, A Renaissance?
Andy Roberts, Duncan Farrant and Philip Sidebottom, Glaxo Wellcome

Liquid Phase Combinatorial Chemistry and Characterization of Intermediates
by Routine Solution State NMR.
Ron Kim, Mahua Manna, Steve Hutchins and Kevin Chapman, Merck & Company

Experimental Aspects of Advanced Diffusion Measurements with PFG NMR.
Donghui Wu, Aidi Chem and Charles S. Johnson, Jr., University of North
Carolina

Evaluation of HTS Probes and Preliminary Results for Biomacromolecular,
Metabolite and Natural Product Compounds.
 R. Andrew Byrd and Siddhartha Sarma, NCI-FCRDC
John Shockcor, Glaxo Wellcome
Gary Martin, Pharmacia & Upjohn, Inc.
Ron Crouch and Toby Zens, Nalorac Corporation

Structure Determination of Proteins in the 30 kD Range and New Methods for
Determining Long Range Order.
G. Marius Clore, National Institutes of Health

High Field Offers More Than High Resolution and Sensitivity.
Ad Bax, National Institutes of Health

Quadruple Resonance Probes - New Tools for Biological NMR.
Arthur Pardi, University of Colorado
Gershon Wolfe and Brian Marsden, Nalorac Corporation

Advances in Solution State NMR Probe Performance.
Toby Zens and Gershon Wolfe, Nalorac Corporation

Application of 1H/31P/X Triple Resonance Experiments in Organometallic
Chemistry.
Emilio Bunel, E.I. DuPont

1H/13C/29Si Triple Resonance Heteronuclear 3D NMR of Organosilicon
Compounds at Natural Abundance.
Peter Rinaldi, University of Akron

Please RSVP by contacting Chris Tierney, Nalorac s ENC Coordinator at
NALORAC, 841-A  Arnold Drive, Martinez, CA 94553.  Phone:  (510) 229-3501
Fax: (510) 229-1651, Email:  christierney@nalorac.com


From owner-structural-nmr@net.bio.net Mon Jan 27 22:00:00 1997
Path: biosci!lhc.nlm.nih.gov!biosci!Merck.Com!yves_aubin
From: yves_aubin@Merck.Com (Yves Aubin)
Newsgroups: bionet.structural-nmr
Subject: Shielding
Date: 27 Jan 1997 14:50:22 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 52
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <199701272104.QAA16919@igw2>
NNTP-Posting-Host: net.bio.net

Dear netters,

Several months ago, there was an inquiry from Istvan about shielding monitors
from the stray fields of NMR magnets.   In the solution I proposed (a homebuilt
shield made of plywood and M6 shielding material) I mentionned the name of
the company from where we got the material but not the person who helped
us.  I guess I wanted to shield this person from too many phone calls...
In fact,
I apologize because I should have acknowledged his help in choosing the right
material for our needs.

This person is named Vahe Ohanian.  Vahe is an engineer at BCL Magnetics and
was very helpful in helping me getting a good shielding material for an
affordable
price (compare to NUmetal).  Talking with him recently,  he was telling me
that he
is testing a new software that will enable him to model magnetic fields and
allow
him to do better suggestions regarding shielding design and shielding material.
He told me that he would be keen in helping people in those issues.  Is
coordinates
are:

			Vahe Ohanian
			BCL-Magnetics
			5045 North Service Rd
			Burlington, ON, Canada,
			L7L 5H6
			phone: (905) 335-2530
			FAX:    (905) 335-8474
			email: magnetic@world.com

I thought that it would be useful to the NMR community to know someone who
knows
about magnetic shielding.

		Hope this helps!


			Yves


Yves Aubin Ph.D.
Merck Frosst Canada Inc.
P.O. Box1005
Pointe-Claire--Dorval
QC, H9R 4P8
tel: (514) 428-3931
fax: (514) 428-8615
email: yves_aubin@merck.com



From owner-structural-nmr@net.bio.net Mon Jan 27 22:00:00 1997
Path: biosci!rutgers!gatech!arclight.uoregon.edu!news.sprintlink.net!news-peer.sprintlink.net!howland.erols.net!newspump.sol.net!ddsw1!news.mcs.net!not-for-mail
From: W R Shefte <tpsi@mcs.net>
Newsgroups: bionet.structural-nmr,sci.materials,sci.med,sci.med.physics,sci.med.radiology
Subject: Re: Monel
Date: Tue, 28 Jan 1997 14:10:48 -0600
Organization: Thermal Processing Solutions, Inc.
Lines: 14
Message-ID: <32EE5D48.737A@mcs.net>
References: <32EC8210.2388@lkeb.medfac.leidenuniv.nl>
NNTP-Posting-Host: tpsi.pr.mcs.net
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.0 (WinNT; I)
Xref: biosci bionet.structural-nmr:1722 sci.materials:21161 sci.med:148918 sci.med.physics:5112 sci.med.radiology:8294

Aart Spilt wrote:
> 
> Hello,
> 
> I'm trying to find out anything about the magnetic characteristics of
> Monel. It is an Alloy of Ni, C, Mn Fe, S, Si and Cu. Please give me all
> the information you have.
> 
> I will be very thankful with every kind of information.
> 
> Aart Spilt
Call Huntington (INCO) Alloys in Huntington WV....(800)334-4626 may
work...otherwise check directory assistance... They are a mfg of monel
inconel and incoly........

From owner-structural-nmr@net.bio.net Mon Jan 27 22:00:00 1997
Path: biosci!agate!howland.erols.net!worldnet.att.net!newsadm
From: Stella Pankey <MonaL@worldnet.att.net>
Newsgroups: bionet.structural-nmr,sci.materials,sci.med,sci.med.physics,sci.med.radiology
Subject: Re: Monel
Date: 28 Jan 1997 00:36:30 GMT
Organization: AT&T WorldNet Services
Lines: 6
Message-ID: <5cjhme$2ub@mtinsc05.worldnet.att.net>
References: <32EC8210.2388@lkeb.medfac.leidenuniv.nl>
NNTP-Posting-Host: 207.146.160.99
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 1.22ATT (Windows; U; 16bit)
Xref: biosci bionet.structural-nmr:1721 sci.materials:21132 sci.med:148793 sci.med.physics:5111 sci.med.radiology:8290

I don't know where you live, but have you tried calling a steel 
distributor?  I am certain a major steel distributor-( look thru the 
industrial yellow pages) can help you. 




From owner-structural-nmr@net.bio.net Tue Jan 28 22:00:00 1997
Path: biosci!agate!news.Stanford.EDU!newshub.internex.net!masters0.news.internex.net!usenet
From: "Woodrow W. Conover" <woody@acornnmr.com>
Newsgroups: bionet.structural-nmr
Subject: Re: (none)
Date: 29 Jan 1997 18:09:34 GMT
Organization: Acorn NMR
Lines: 42
Message-ID: <01bc0e0f$c834e7e0$3de29fce@almond>
References: <Pine.SGI.3.93.970129111944.15885A-100000@caribe.chem.uky.edu>
NNTP-Posting-Host: 206.159.226.61
X-Newsreader: Microsoft Internet News 4.70.1160

When the NMR shims behave differently with a change in the sample it is
possible that the sample has some solid materials in the solution.  It is
not unusal for the solid material to be iron particles which can really
mess up the shimming. Examine the sample with a magnifying glass under a
strong light and see if you have some solid stuff in the solution.

If the solution is clean, it could be a defective NMR tube but this would
be very unusal.  I suggest:

1. Go back to a known sample and confirm that it is working okay.  If it is
not then something happened to the NMR system (wrong shim values, bad shim
power supply, broken NMR probe and etc).
2. Clean the NMR tube and filter the sample through glass wool into the
cleaned NMR tube.  Make sure there is about the sample volume of sample as
in the known good sample, because end effects of a short sample can make
shimming difficult.
3.  If this fails try another tube.

woody@acornnmr.com
  
Arthur Cammers <cammers@CARIBE.CHEM.UKY.EDU> wrote in article
<Pine.SGI.3.93.970129111944.15885A-100000@caribe.chem.uky.edu>...
> 	I was shimming the magnet (300MHz widebore) the other day with an
> unknown in the probe.  The unknown I ran just before I put this one in
the
> probe gave me beautiful line shapes with widths at half height of ~0.5Hz,
> 1H. 
> 	With this sample maximizing the lock signal with the spinning
> shimming parameters did not optimize the peak shape.  I suspect a bad
> tube.  Is this behavior a sign of a untrue nmr tube?  Could someone else
> with experience with this aberrant shimming behavior please comment?
> Thanks.
> 
> 
> Arthur Cammers-Goodwin				
> Assistant Professor of Chemistry		
> Department of Chemistry-Organic Division 	
> University of Kentucky				
> acgood1@pop.uky.edu
> http://www.chem.uky.edu/research/cammers/cammerscv.html
> 
> 

From owner-structural-nmr@net.bio.net Tue Jan 28 22:00:00 1997
Path: biosci!CARIBE.CHEM.UKY.EDU!cammers
From: cammers@CARIBE.CHEM.UKY.EDU (Arthur Cammers)
Newsgroups: bionet.structural-nmr
Subject: (none)
Date: 29 Jan 1997 08:23:43 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 18
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.93.970129111944.15885A-100000@caribe.chem.uky.edu>
NNTP-Posting-Host: net.bio.net

	I was shimming the magnet (300MHz widebore) the other day with an
unknown in the probe.  The unknown I ran just before I put this one in the
probe gave me beautiful line shapes with widths at half height of ~0.5Hz,
1H. 
	With this sample maximizing the lock signal with the spinning
shimming parameters did not optimize the peak shape.  I suspect a bad
tube.  Is this behavior a sign of a untrue nmr tube?  Could someone else
with experience with this aberrant shimming behavior please comment?
Thanks.


Arthur Cammers-Goodwin				
Assistant Professor of Chemistry		
Department of Chemistry-Organic Division 	
University of Kentucky				
acgood1@pop.uky.edu
http://www.chem.uky.edu/research/cammers/cammerscv.html


From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!agate!howland.erols.net!feed1.news.erols.com!news.bbnplanet.com!cpk-news-hub1.bbnplanet.com!news.mindspring.com!uunet!in1.uu.net!138.133.17.7!amgen!usenet
From: Thomas J Hoeffel <thoeffel@amgen.com>
Newsgroups: bionet.structural-nmr
Subject: position available
Date: Wed, 29 Jan 1997 16:35:33 -0800
Organization: Amgen
Lines: 36
Message-ID: <32EFECD5.2781@amgen.com>
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.0 (X11; I; IRIX 6.2 IP22)

Amgen, the world's largest biotechnology company, seeks the 
computational/analytical talents of an experienced UNIX System
Administrator to support our Research Staff.

In collaboration with current system administration staff the
incumbent will be responsible for administration of the installed
UNIX systems in Research with a focus on SGI systems and scientific 
applications in a graphical environment. In addition the incumbent 
will be responsible for  recommending additonal equipment purchases, 
participating in Disaster Recovery Planning/Testing and assisting 
the  strategic planning for the use of computers in Research.

This position requires a BS/MS degree in the Life Sciences or 
Computer Science or the equivalent plus a minimum of three years 
of UNIX System Administration experience with a minimum of two 
years of IRIX experience. Familiarity with the Research Computing 
Environment and the ability to work in multi-disciplinary teams is 
a must. Additional experience with other UNIX systems (HP, SUN, 
DEC Alpha) is highly desirable, experience with Macintosh and/or 
Wintel PC's as well as VMS would be valuable.

We value and encourage the diverse perspectives essential to the 
process of discovery.

At Amgen, your accomplishments will be rewarded, our generous 
compensation and benefits package includes a retirement and savings 
plan, an on-site fitness center and three weeks vacation. For 
immediate consideration, mail your resume to: Amgen, Staffing 
Job Code:  T10139, P.O. Box 2569, Thousand Oaks, CA 91319-2569. 
E-mail: jobs@amgen.com. Please consult our on line Job Bulletin Board 
at http//amgen.bio.com for information on other career opportunities 
available at Amgen. Principals only, please.

EEO/AA Employer M/F/D/V

--
*** Disclaimer: These are the opinions of the poster not Amgen Inc.***

From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!OTTER.BIOCHEM.UBC.CA!mcintosh
From: mcintosh@OTTER.BIOCHEM.UBC.CA (Lawrence McIntosh)
Newsgroups: bionet.structural-nmr
Subject: color printers
Date: 29 Jan 1997 17:28:43 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 31
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.91.970129171001.12100H-100000@otter.biochem.ubc.ca>
NNTP-Posting-Host: net.bio.net


Hello:

I am interested in purchasing a color printer
to make (ideally inexpensive and "routine") color 
prints of NMR spectra, structures, etc. Certainly a 
Codonex dye sublimination printer is the Cadillac, but not 
cheap. 

Does anyone have experience with color laser
printers or  "bubble-jet" printers in terms
of computers (SGI or Sun), NMR programs (Felix, NMRdraw),
and structure programs (postscript, hpgl, etc)?

Thanks!!

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

Lawrence McIntosh
2146 Health Sciences Mall
Departments of Biochemistry and Chemistry
University of British Columbia
Vancouver, BC, Canada V6T 1Z3

mcintosh@otter.biochem.ubc.ca
ph:  (604) 822-3341
fax: (604) 822-5227
http://www.biochem.ubc.ca/

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~


From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!MANGO.UOREGON.EDU!strain
From: strain@MANGO.UOREGON.EDU (Michael Strain)
Newsgroups: bionet.structural-nmr
Subject: Re: Tube Problem?
Date: 30 Jan 1997 09:55:59 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 93
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.91.970130094317.1484F-100000@mango.uoregon.edu>
References: <Pine.SGI.3.95.970130074100.25495A-100000@mole.bio.cam.ac.uk>
NNTP-Posting-Host: net.bio.net


A few years ago Wilmad sent out test "turbines", a length of heavy-wall
precision glass tubing into which a 5mm NMR tube could be inserted and spun
by hand.  Good tubes would spin easily for many seconds after an initial
twist.  Tubes with poor camber or concentricity would not spin well.  I 
don't know if these test turbines are still available. 

Another way to check for a bad tube is using the real-time swept tuning 
display available on some spectrometers with the sample in the magnet.
With the sample spinning, the tuning dip should be quite stable, 
moving side-to-side only a little.  I have seen tubes so bad that the dip 
moved about like a prairie cyclone.  Besides yielding poor NMR 
performance, such tubes also have the potential to damage the coil insert 
in the probe.  

--Mike


-----------------------------------------------------------------------
Michael Strain				strain@mango.uoregon.edu
Institute of Molecular Biology		desk/voice-mail: 541-346-4605	
   and Department of Chemistry		lab:		 541-346-4036	
University of Oregon			FAX:		 541-346-5891
Eugene, OR 97403

NMR instrument rooms:  600MHz, 6-5161;  500MHz, 6-4621;  300MHz, 6-4616
-----------------------------------------------------------------------


On 30 Jan 1997, Mark J Howard wrote:

> 
> 
> 
> 
> > 	I was shimming the magnet (300MHz widebore) the other day with an
> > unknown in the probe.  The unknown I ran just before I put this one in the
> > probe gave me beautiful line shapes with widths at half height of ~0.5Hz,
> > 1H. 
> > 	With this sample maximizing the lock signal with the spinning
> > shimming parameters did not optimize the peak shape.  I suspect a bad
> > tube.  Is this behavior a sign of a untrue nmr tube?  Could someone else
> > with experience with this aberrant shimming behavior please comment?
> > Thanks.
> 
> I would agree with Woodrow Conover and acertain whether the fault
> definately lies with the sample or spectrometer.  As an alternative to
> using glass wool to filter your sample, you can get microfilters from
> Millipore or Whatman which will filter down past 0.2um.  There are
> filters specially designed for coping with small sample volumes without
> excessive loss due to liquid absorption into the filter.
> 
> On the point of faulty NMR tubes, I have seen this before.  I had a
> protein sample which just would not shim well and as a result the line
> shape and presaturation was terrible.  I checked the spectrometer with a
> sample that I knew was ok.  So I transfered the original to a second tube
> and all my problems were cured.  I did take a very close look at the
> 'faulty'NMR tube and I found a very small (~0.2mm) scratch on the outside  
> surface of the tube just at the position where the probe coils would be,
> so I put the problem down to that.  However, I then had problems with the
> rest of the tubes that came boxed with the original problem tube.  They
> all would not shim well etc.., but when samples were placed in tubes from
> another batch or I knew were ok, they were fine.  I made sure nothing
> had been accidently introduced during the washing process before they were 
> used, but they still were not shimming well.  I could not see any
> visible fault on the other tubes unlike the one that originally caused
> the problem.  In the end, I sent the box back to the manufacturer and got
> a new box sent free so I put the experience down to a bad batch of
> tubes.  So if you do suspect a tube of being faulty, it may be wise to
> isolate all tubes from the same box and try a tube you know to be sound if
> the problem occurs with tubes from the same batch.
> 
> 
> Mark Howard
> 
> ********************************************************
> Dr. Mark Howard
> Protein NMR Spectroscopy                    |
> Dept. of Biochemistry                       | 
> University of Cambridge                    || |
> Tennis Court Road	      |		   || |    |
> Cambridge, UK.		     ||	        | ||| |    |
> CB2 1QW   		____|||||___/\__||||||||___|_|_ 
> 
> Tel: (01223) 333662  Fax: (01223) 333661
> E-mail: mjh2@mole.bio.cam.ac.uk
> ********************************************************
> 
> 
> 
> 
> 
> 

From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!PHOENIX.PRINCETON.EDU!ipelczer
From: ipelczer@PHOENIX.PRINCETON.EDU (Istvan Pelczer)
Newsgroups: bionet.structural-nmr
Subject: Re: color printers/NMR programs
Date: 30 Jan 1997 05:48:43 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 24
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SUN.3.91.970130083053.25140A@phoenix.princeton.edu>
References: <Pine.SGI.3.91.970129171001.12100H-100000@otter.biochem.ubc.ca>
NNTP-Posting-Host: net.bio.net



This comment is not about color printers in particular, only a note on 
the sideline about NMR programs mentioned by Lawrence ("...NMR programs 
(Felix, NMRdraw)...").
Beside Felix and nmrPipe/nmrDraw I can truly recommed NMRView for 
visualization and spectrum analysis, which complements the other two and 
takes direct input from each.  In addition, it generates plots both in 
postscript and hpgl format now.  It is a free software.
Whoever is interested please, take a look at:
	 http://www.ges.com/nmrview/
or contact Bruce Johnson (bruce_johnson@merck.com).
All the best,

Istvan

wwww,wwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwwww
Istvan Pelczer, Ph.D.		       		Email: ipelczer@princeton.edu
Senior NMR Spectroscopist
Princeton University
Department of Chemistry, Frick Lab.,	 		 ph#  (609) 258 2342
Washington Road						fax#  (609) 258 6746
Princeton,  NJ 08544, USA


From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!rutgers!gatech!news.sprintlink.net!news-peer.sprintlink.net!arclight.uoregon.edu!news.bbnplanet.com!su-news-hub1.bbnplanet.com!csn!nntp-xfer-1.csn.net!csn!nntp-xfer-2.csn.net!tali.UCHSC.edu!chris
From: "Joel Guthridge, Ph.D." <Joel.Guthridge@uchsc.edu>
Newsgroups: bionet.structural-nmr
Subject: Isotopic labeling of proteins produced in Pichia
Date: Thu, 30 Jan 1997 09:43:35 -0700
Organization: Div. of Rheumatology, Univ. of Colorado Health Sci. Ctr.
Lines: 13
Message-ID: <32F0CFB7.23AC@uchsc.edu>
Reply-To: Joel.Guthridge@uchsc.edu
NNTP-Posting-Host: handel.uchsc.edu
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.0Gold (Win95; I)

Does anyone have some information on isotopic labeling of proteins
produced in Pichia?  I'm looking for both some protocols to use as
starting points and information on sources of labeled reagents that
will work well in the methanol induction of protein expression in
Pichia.

Thanks,

Joel Guthridge
Div. of Rheumatology,
Univ. of Colorado Health Sci. Ctr.
Denver, CO
email:  Joel.Guthridge@uchsc.edu

From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!rutgers!gatech!arclight.uoregon.edu!su-news-hub1.bbnplanet.com!news.bbnplanet.com!cpk-news-hub1.bbnplanet.com!news.maxwell.syr.edu!newsfeeds.sol.net!nntp.uio.no!news.kth.se!tybalt.admin.kth.se!celsiustech.se!seunet!news2.swip.net!mailgate.astra.com!usenet
From: Gunnar Gronberg <gunnar.gronberg@draco.se.astra.com>
Newsgroups: bionet.structural-nmr
Subject: Position available
Date: Thu, 30 Jan 1997 11:36:03 -0800
Organization: Astra Draco AB
Lines: 40
Message-ID: <32F0F823.2AEC@draco.se.astra.com>
NNTP-Posting-Host: dpggg.draco.se.astra.com
Mime-Version: 1.0
Content-Type: text/plain; charset=iso-8859-1
Content-Transfer-Encoding: 8bit
X-Mailer: Mozilla 2.0 (Win16; I)

NMR Specialist 
Department of Medicinal Chemistry 
ASTRA DRACO
Lund, Sweden

The Department of Medicinal Chemistry is involved in the creation of new 
drugs. New chemical compounds are synthesised and their structures 
characterised. Properties essential to the biological effects are 
investigated and optimised. The department consits of a number of  
organic chemistry groups orientated towards projects in different phases 
of  development together with special competence in the areas of 
Combinatorial Chemistry, Radio-chemistry, NMR, Computational Chemistry 
and Information Technology/System Management. The NMR facilitiy consists 
three high resolution spectrometers: 300, 400 and 500 MHz (Varian).

The Department of Medicinal Chemistry invites applications for a position 
as an NMR Chemist to perform structural/conformational as well as 
physical chemistry related NMR investigations of drug molecules and their 
interactions with target molecules and pharmaceutical preparations. The 
primary interactions will be with Medicinal Chemists and Computational 
Chemists but also with other disciplines within ASTRA DRACO. Other 
responsibilities include maintenance of the NMR systems, development of 
automation routines as well as  training and consultation with chemists. 

Candidates are expected to posses a PhD in Physical Chemistry or Organic 
Chemistry with NMR experience from peptides, bio membranes,  protein 
ligand interactions, diffusion measurements etc. Experience of UNIX 
administration, pulse programming and maintenance of NMR systems are 
important qualifications.

 
Further information is available directly from:
Gunnar Grönberg +46 46 33 70 65 
email: gunnar.gronberg@draco.se.astra.com 
http://www.astra.com
Please submit application, curriculum vitae and  letters of 
recommendation before February 10, 1996 to: 

Personalavdelningen, 
Lena Ask, Astra Draco AB, Box 34, 221 00 Lund, Sweden

From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!agate!howland.erols.net!surfnet.nl!swidir.switch.ch!scsing.switch.ch!elna.ethz.ch!usenet
From: Reto Koradi <kor@mol.biol.ethz.ch>
Newsgroups: bionet.structural-nmr
Subject: Re: color printers
Date: Thu, 30 Jan 1997 11:18:00 +0100
Organization: Swiss Federal Institute of Technology (ETHZ)
Lines: 36
Message-ID: <32F07558.4346@mol.biol.ethz.ch>
References: <Pine.SGI.3.91.970129171001.12100H-100000@otter.biochem.ubc.ca>
NNTP-Posting-Host: odin.ethz.ch
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
X-Mailer: Mozilla 3.0Gold (X11; I; SunOS 5.4 sun4d)
To: Lawrence McIntosh <mcintosh@OTTER.BIOCHEM.UBC.CA>

Lawrence McIntosh wrote:
> Does anyone have experience with color laser
> printers or  "bubble-jet" printers in terms
> of computers (SGI or Sun), NMR programs (Felix, NMRdraw),
> and structure programs (postscript, hpgl, etc)?

We frequently use Canon color laser printers (CLC 500), and
they are excellent. But these are large units, suited for
central printing facilities.

We also have an Apple Color LaserWriter 12/600 PS available.
The output quality is fine, but it happens that it refuses to
print PostScript files that other printers accept. It also
completely broke down once. These color laser printers are
getting cheaper, you can probably get something in that class
for around $6000. Tektronix has an excellent reputation for
their color printers, in a similar price class.

About ink jet printers, I haven't seen any recently. But I'm
told that they have improved quite dramatically, and have very
good output quality. They're typically slow. Good brands there
are Epson and HP.

With all color printers, you should not only look at the price
for buying it, but also the costs per page. Ink jet printers
need frequent ink refills, and might need special paper for
good results. Laser printers eat toners. All these things tend
to be quite expensive.

The more expensive (laser) printers will probably come with
PostScript support, and that's what most programs produce.
Ink jet printers normally don't understand PostScript, or only
as an added option. You can either go the free (Ghostscript)
way, or pay for commercial solutions, like Impressario by SGI.
-- 
Reto Koradi (kor@mol.biol.ethz.ch, http://www.mol.biol.ethz.ch/~kor)

From owner-structural-nmr@net.bio.net Wed Jan 29 22:00:00 1997
Path: biosci!MOLE.BIO.CAM.AC.UK!mjh2
From: mjh2@MOLE.BIO.CAM.AC.UK (Mark J Howard)
Newsgroups: bionet.structural-nmr
Subject: Tube Problem?
Date: 30 Jan 1997 00:08:11 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 61
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <Pine.SGI.3.95.970130074100.25495A-100000@mole.bio.cam.ac.uk>
NNTP-Posting-Host: net.bio.net





> 	I was shimming the magnet (300MHz widebore) the other day with an
> unknown in the probe.  The unknown I ran just before I put this one in the
> probe gave me beautiful line shapes with widths at half height of ~0.5Hz,
> 1H. 
> 	With this sample maximizing the lock signal with the spinning
> shimming parameters did not optimize the peak shape.  I suspect a bad
> tube.  Is this behavior a sign of a untrue nmr tube?  Could someone else
> with experience with this aberrant shimming behavior please comment?
> Thanks.

I would agree with Woodrow Conover and acertain whether the fault
definately lies with the sample or spectrometer.  As an alternative to
using glass wool to filter your sample, you can get microfilters from
Millipore or Whatman which will filter down past 0.2um.  There are
filters specially designed for coping with small sample volumes without
excessive loss due to liquid absorption into the filter.

On the point of faulty NMR tubes, I have seen this before.  I had a
protein sample which just would not shim well and as a result the line
shape and presaturation was terrible.  I checked the spectrometer with a
sample that I knew was ok.  So I transfered the original to a second tube
and all my problems were cured.  I did take a very close look at the
'faulty'NMR tube and I found a very small (~0.2mm) scratch on the outside  
surface of the tube just at the position where the probe coils would be,
so I put the problem down to that.  However, I then had problems with the
rest of the tubes that came boxed with the original problem tube.  They
all would not shim well etc.., but when samples were placed in tubes from
another batch or I knew were ok, they were fine.  I made sure nothing
had been accidently introduced during the washing process before they were 
used, but they still were not shimming well.  I could not see any
visible fault on the other tubes unlike the one that originally caused
the problem.  In the end, I sent the box back to the manufacturer and got
a new box sent free so I put the experience down to a bad batch of
tubes.  So if you do suspect a tube of being faulty, it may be wise to
isolate all tubes from the same box and try a tube you know to be sound if
the problem occurs with tubes from the same batch.


Mark Howard

********************************************************
Dr. Mark Howard
Protein NMR Spectroscopy                    |
Dept. of Biochemistry                       | 
University of Cambridge                    || |
Tennis Court Road	      |		   || |    |
Cambridge, UK.		     ||	        | ||| |    |
CB2 1QW   		____|||||___/\__||||||||___|_|_ 

Tel: (01223) 333662  Fax: (01223) 333661
E-mail: mjh2@mole.bio.cam.ac.uk
********************************************************






From owner-structural-nmr@net.bio.net Thu Jan 30 22:00:00 1997
Path: biosci!agate!newsgate.cuhk.edu.hk!news.glink.net.hk!uunet!in2.uu.net!192.89.123.24!nntp.inet.fi!news.csc.fi!ousrvr3.oulu.fi!ernst.oulu.fi!not-for-mail
From: eha@ernst.oulu.fi (Esa Haapaniemi)
Newsgroups: bionet.structural-nmr
Subject: Re: color printers
Date: 31 Jan 1997 12:50:26 GMT
Organization: University of Oulu
Lines: 20
Distribution: world
Message-ID: <5cspqi$48j@ousrvr3.oulu.fi>
References: <Pine.SGI.3.91.970129171001.12100H-100000@otter.biochem.ubc.ca>
NNTP-Posting-Host: ernst.oulu.fi
Mime-Version: 1.0
Content-Type: text/plain; charset=iso-8859-1
Content-Transfer-Encoding: 8bit
X-Newsreader: TIN [UNIX 1.3 950824BETA PL0]

Lawrence McIntosh (mcintosh@OTTER.BIOCHEM.UBC.CA) wrote:
: Does anyone have experience with color laser
: printers or  "bubble-jet" printers in terms
: of computers (SGI or Sun), NMR programs (Felix, NMRdraw),
: and structure programs (postscript, hpgl, etc)?

Yes and no. We have used several B&W and colour printers here, but
normally they are connected to PCs. Worst problem with getting them
to work on workstations is the need of drivers. So in our department
files are moved to PC and printed from there (in need). With that
we can use all kinds of possible printers/plotters/diaprinters/...

BTW Epson is currently working on 1400 dpi colour bubblejet that
should be out on March. The Pro model should handle postscript and
network printing and might be suitable for workstations as well.

	Esa Haapaniemi
	University of Oulu
	Department of Chemistry
	Finland

From owner-structural-nmr@net.bio.net Thu Jan 30 22:00:00 1997
Path: biosci!SCRIPPS.EDU!georges
From: georges@SCRIPPS.EDU (Georges Mer)
Newsgroups: bionet.structural-nmr
Subject: subscribe
Date: 30 Jan 1997 18:15:12 -0800
Organization: BIOSCI International Newsgroups for Molecular Biology
Lines: 1
Sender: daemon@net.bio.net
Distribution: world
Message-ID: <32F154E9.41C67EA6@scripps.edu>
NNTP-Posting-Host: net.bio.net

subscribe

From owner-structural-nmr@net.bio.net Fri Jan 31 22:00:00 1997
Path: biosci!agate!nntpfeed.doc.ic.ac.uk!sunsite.doc.ic.ac.uk!warwick!lyra.csx.cam.ac.uk!bioc.cam.ac.uk!wb104
From: wb104@bioc.cam.ac.uk (Wayne Boucher)
Newsgroups: bionet.structural-nmr,sci.techniques.mag-resonance
Subject: Azara 2.0 release
Date: 1 Feb 1997 10:33:27 GMT
Organization: Somewhere in the University of Cambridge
Lines: 15
Distribution: world
Message-ID: <5cv65n$lro@lyra.csx.cam.ac.uk>
NNTP-Posting-Host: giza.bioc.cam.ac.uk
Xref: biosci bionet.structural-nmr:1739 sci.techniques.mag-resonance:1966

Azara (v2.0) is a suite of programs to process and view NMR data,
including 1D, 2D and 3D maximum entropy processing (in an ND data set).

The distribution can be obtained via anonymous ftp at ftp.bio.cam.ac.uk
in the directory pub/azara, or via the Web at www.bio.cam.ac.uk/azara.

The distribution comes with either the source code plus executables
or with just the executables (compiled for Silicon Graphics machines).

Several of the programs have been designed so that they may be used in
conjunction with Per Kraulis' program Ansig.  In particular, the data
structures are consistent between Ansig and Azara.

The programs were developed initially at, and now in conjunction with,